How can one improve linearity of an RTD based temperature
measurement system?
Answers were Sorted based on User's Feedback
Answer / guest
RTD has +/- 0.4% nonlinearity when used over 0 to 200 degC.
This is due to the relationship R= R0*(1+aT+bT^2). The
change in resistance for 0 to 100 degC is 38.5ohms for a
Pt100, while it is lesser at 37.3 ohms fro 100 to 200 degC.
Using a two slope method, one for 0 to 100 deg and another
for 100 to 200 degC, the linearity error can be brought
down from +/- 4% to +/- 0.1%.
| Is This Answer Correct ? | 5 Yes | 3 No |
what is exact time to study dcs and plc cources
What is the start up procedure of desalter
what is the calibration procedure for magnetic flow meter @ onsite
Why we not use 2-20Ma or 1-20Ma
can u please tell me the different types of valve uses and wt purposes it is used ~!(controlling the flow other than that which valve should be used in what place)
protection relay watchdog contact operation?
What is mean by Ethernet, Device net and Control net? For what application we are using these?
What is the Flow rate ?
what is difference between differial pressure & delta pressure?
why we are using magnet for calibration of gas detector?
Yokogawa mass flow meter shows negative reading...what r the main causes for this problem?...
Why Pressure Transmitter is NOT installed before up-stream of an ORIFICE PLATE? in a flow measurement.
Civil Engineering (5086)
Mechanical Engineering (4453)
Electrical Engineering (16638)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)