whate is difference between calibration and validation
Answers were Sorted based on User's Feedback
Answer / siddhesh
Calibration is a tool to ensure that the instrument or equipment is working well & Validation is a tool to ensure that the analytical method is working well.
Is This Answer Correct ? | 65 Yes | 7 No |
Answer / thiruathi swamy
Validation mean documented evidence that a particular
instrument or process produce results meet predetermined
specification.
Calibration means particular instrument or device produce
results by comparison of known reference standard.
Is This Answer Correct ? | 39 Yes | 3 No |
Answer / ragav
Validation in the pharmaceutical and medical device
industry is defined as the documented act of demonstrating
that a procedure, process, and activity will consistently
lead to the expected results.
Making measurements with any analytical method or
instrument requires calibration to ensure the accuracy of
the measurement. There are two common calibration
procedures: using a working curve, and the standard-
addition method. Both of these methods require one or more
standards of known composition to calibrate the
measurement.
Is This Answer Correct ? | 23 Yes | 9 No |
will calibratin only for instruments ,. but we can perform
validation not only for instuments for method allso,,..
calibratiom = checking the instrument with standards
validation = validating the existing method or instrument
Is This Answer Correct ? | 13 Yes | 0 No |
Answer / arvind singh yadav
Validation is a documented program that provides high degree of assurance that a specific process, equipment, method or system consistently produces a result meeting pre-determined acceptance criteria.Shall be performed as per validation protocol.
Calibration is a demonstration that, a particular Instrument or device produces results with in specified limits by comparisons with those produced by a reference or traceable standard over an appropriate range of measurements.Shall be performed as per calibration SOP.
Is This Answer Correct ? | 11 Yes | 1 No |
Answer / chandra bhushan tiwari
Calibration is done by known device (standard)againt the
the unknown device, it means instrument are working
properly or not and can be adjusted within tolerance.
Validation is to ensure the method which is to be used for
its intended purpose or not.
Is This Answer Correct ? | 10 Yes | 1 No |
Answer / nirav panchal
Calibration is done to nulify or remove the deviation by comparing with known or reference standard,
while Validation is done to prove that Consistency of expacted results of Process,Method,or Procedure
Is This Answer Correct ? | 6 Yes | 2 No |
calibration:is to check instrument working condition with standards
validation: is to make shore that instrument and methods and processare consistently producing repeatable results as per in house requirement .
calibration gives assurance about only instrument but validation gives assurance about method, and process and man instruments
Is This Answer Correct ? | 1 Yes | 0 No |
Answer / nimesh trivedi
Calibration means system performance checking
Validations means Mfg. Process or analytical method performance checking
Is This Answer Correct ? | 0 Yes | 0 No |
please explain about aluminium hydroxide assay
Why perform dissolution multimedia study in dissolution Development
How can the GC or HPLC method is selected to determine the impurity profile in drug product?
What is Formula of Potency calculation on Dried basis, As is basis& Anhydrous basis.
what is diffrence between Specific optical rotation, optical rotation?
what is the defination of calibration & verification?
what is %labelled amount in content uniformity of dosage unit and its calculation?
What is the defination of Bulk density?
why does the result variation in hplc and gc with same sample ?
In which cases factor of KF Reaget got higher side
What is difference between known and unknown impurities
How to caluculate the RRF and correction factor in hplc
1 Answers Alkem Labs, Aurobindo, Rakshit Drugs, SDLPL,