In specificity what parametrs are consider why that parametrs
Answers were Sorted based on User's Feedback
Answer / sameer
check peak purity of main peak nd known impurity peak.check interference of Main/Known impurity peak in diluent,blank or in placebo
Is This Answer Correct ? | 2 Yes | 0 No |
Answer / manojmishra
The test specificity in analytical method validation is
considred as the identification test.
In this test we considred suitable identification of
substance to be validated as retention time of substance
peak in HPLC and GC, Wavelength in UV, and some suitable
method as colour test or RF value for chemical and TLC
validation.
Therefore, you can consider, %RSD of peak RT,Resolution
between the peak as system suitability etc.
Is This Answer Correct ? | 0 Yes | 0 No |
What is the calibration of uv process and preparation
HOW RRF IS CALCULATED IN HPLC
What is basic difference between method validation,process validation
For limit test of heavy metals in BP, Method C require that the substance is ignited at a temperature not exceeding 800 °C. Why confines such the temperature?
what is the use of performing tapped density in pharmaceutical industry
if you get peak in blank then what require to do?
For LOD and LOQ determination which method is accurate? (slopemethod or s/n ratio method)
stability study is going on up to 3years for famotidine-USP but the pharmacopeia is revised and tlc test is replaced with hplc test. At the 4th year frequency the product is failed in hplc test .how can i assain the expiry or retest date and can i stop or continue the programme?
Why only lead standard solution is uesd to perform heavy metal test.
what is the difference between IR and FTIR instrument analysis
Why using Holmium oxide in UV calibration
8 Answers Dr Reddys, Matrix, Nakoda Chemicals, Orchid,
column washing procedure for normal phase and reverse phase?