why we use only vanilline , acetanilide,& caffeine For the
Calibration of melting point appartus?
Answers were Sorted based on User's Feedback
Answer / sachin ghogare
These chemicals are very stable for M.P. test and covers
range between 0-300°c .Easily available in market .
| Is This Answer Correct ? | 50 Yes | 4 No |
These chemicals are very stable and covers
range between 0-300°c .Easily available in market .
vanilline - 81°C to 83°C
Acetanilide - 114°C to 116°C
Phenacetin - 134°C to 136°C
Sulfanilamide - 164.5°C to 166.5°C
Sulfapyridine - 191°C to 193°C
Caffeine - 234°C to 237°C
| Is This Answer Correct ? | 30 Yes | 2 No |
vanilline - 81°C to 83°C (range-1.5°)
Acetanilide - 164°C to 165.7°C (range-1.0°)
Caffeine - 235.6°C to 237.5°C (range-1.0°)
All three standards starting & Ending points is within 1.5°
only.
In our operational way also it covers minimum, middle &
Maximum calibration covered.
| Is This Answer Correct ? | 19 Yes | 4 No |
Answer / ajay vishwakarma
These standard have very sharp melting point in their range
and also they are easily available in market.
| Is This Answer Correct ? | 10 Yes | 3 No |
Answer / ishwar
These standards having sharp melting point.stability.
| Is This Answer Correct ? | 5 Yes | 0 No |
Answer / rupesh
This standards give Linearity
& covers range between 0-300°c .Easily available in market.
| Is This Answer Correct ? | 1 Yes | 0 No |
Answer / renu saini
Becoz its covers whole range of melting point 0- 300°c.it is easily avialable in market.all are usp reference standards
| Is This Answer Correct ? | 2 Yes | 1 No |
Wt s the use of linearity in HPLC method validation
what is stability study ? why it is needed and how itis done
why do we use around only 60 mg of K2cr2o7 in the UV calibration -control of absorbance test?
how to pepare 6.8 pH phosphate buffer
I want analyze my compound at 325 nm so which standard shall i use to calibrate my HPLC system? How i can assure my hplc detector is suitable or not?
what is the limit of Correlation for IR test compare to the standard and test
difference between IR & NMR spectroscopy ?
why in uv we take k value less than 1
Why we perform the ir analysis?
Why only lead standard solution is uesd to perform heavy metal test.
If inhouse hplc related substance method is completly diffrent from Usp for finished proďuct with diffrent impurities then how require to prove method equivalecy?
priciple of gc & hplc? (properly describe)
2 Answers Cadila, Cipla, Reddy Labs,