what r the parts calibrated in gc
Answers were Sorted based on User's Feedback
Answer / s.ramanjaneyulu
The GC Calibration parameters are
a)FID(Flame Ionisation Detector) for Responce purpose
b)ECD Electron Capture Detector) for Responce purpose
c)Head-Space sampler
d)H2 Flow,N2 or He Flow (Carrier gas flow rate what u are
using)
| Is This Answer Correct ? | 1 Yes | 1 No |
Answer / surya
1. Flow rate accuracy
2. Oven temperature
3. System precision
4. Injector linearity
5. detector linearity
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / venkatanarayana
Leak test,System Precision , Detector Linearity
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / venkat
system sutabulity
detecter linearty
auto injecter vial position check
injection volume
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / viswanathan
1)Detector flow and column flow
2)Column oven temp minimum and maximum (based on operating range)
3)Detector linearity
4)injector precision and linearity
5)HS vial detection test
6)Noise and drift
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / umamahesh
carrier gas flow check,detector linearity,injection precision by auto sampler and injection precision of head space
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / gangaram kolekar
1.Linearity of detector response ( FID )
2.Temperature Calibration ( Column Oven )
3.Flow rate Accuracy
4.Correlation coefficient
5.Precession
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / gaurav
Auto injector, detectors, most important: Column
| Is This Answer Correct ? | 1 Yes | 2 No |
what is the difference between potentiometric assay and hplc assay
How we will choose pH of a mobile phase if analyte pKa value is 4
how much mass should be there in volumetric flask while in preparation of sample for assay?
what is the origin to prepare standard operating procedure
why 0.1M OR 1M Perchloric acid standardisation done after 24 hours?
If i performing one analysis on C8column then the same analysis done on C18 column then what will happen about Rt;Resulution;Peak shape Explain ?
UV and PDA detector, which have less signal to noise ratio?
explain 5 components of gas chromatography ?
1 Answers College School Exams Tests, Nucon,
Why pH range is maximum 14.0
Why we use Caffeine in hplc calibration
Whatn is HPLC
1 Answers HPLC, Jubilant, Orchid,
Shall i use a same hplc column for two different products.?