what r the parts calibrated in gc
Answers were Sorted based on User's Feedback
Answer / mahesh
Many parts like Auto injector, Detector,Injection volume,
Flow rate,
| Is This Answer Correct ? | 21 Yes | 1 No |
Answer / kuchela rao athota
1)CARRIER FLOW CHECK
2)COLUMN OVEN TEMP..(ISOTHERMAL TEMP.. CHECK)
3)COLUMN OVEN TEMP..(PROGRAMMABLE TEMP..CHECK)
4)VIAL HEATING BLOCK TEMP..
5)TRANSFERLINE TEMP..
6)PRECISION TEST
7)LENIARITY TEST
8)FID SENSIVITY TEST
| Is This Answer Correct ? | 15 Yes | 3 No |
Answer / alpesh patel
Parts like Auto injector, Detector, Oven, HS
| Is This Answer Correct ? | 4 Yes | 1 No |
Answer / jyothi
gas flow rate
carrier gas flow rate
detector gas
oven calibration
detector calibration
these above all calibration parameters in gas
chromatography.
| Is This Answer Correct ? | 4 Yes | 1 No |
Answer / a.ramakrishnan
Parameters for GC Callibration
1. Carrier gas flow rate.
2. Split Flow Callibration.
3. Injector Linearrity
4. Detector Response
5. Column oven temp.
| Is This Answer Correct ? | 3 Yes | 0 No |
Answer / surya
Flow rate , Oven , Auto injector , Detector response , HS
| Is This Answer Correct ? | 3 Yes | 2 No |
Answer / vimal
AUTO INJECTOR
DETECTOR
COLUMN OVEN
VIAL TEMPARATURE
HS
| Is This Answer Correct ? | 2 Yes | 1 No |
Answer / kanakadurga
Calibration parameters in Gc
1.Flow calibration:hydrogen gas,air and carrier gas
2.Column oven temperature:50,100,200,300
3.Injector precesion:6 times
4.Detector linearity:20% to 120%
| Is This Answer Correct ? | 1 Yes | 0 No |
Answer / shanthi
Injector precission
Liearity of the detector
Oven temperature calibration
Carrier gas flow rate calibration
| Is This Answer Correct ? | 1 Yes | 0 No |
Answer / s v ramesh
1. OVEN TEMPERATURE
2. HSS SENSITIVITY WITH USING ETHANOL OR IPA
3. CAPILLARY SENSITIVITY OF FID USING FID SAMPLE
4. DETECTOR LINEARITY.
| Is This Answer Correct ? | 0 Yes | 0 No |
what buffer you can select for acidic nature analyte and how ?
what is Hilic chromatography?
IN pH CALIBRATION,CALIBRATION IS DONE FOR pH METER OR ELECTRODE?HOW WILL YOU JUDGE THAT THE METER TO BE CALIBRATED WITH 4,7,9.2 BUFFERS?EXPLAIN TEMPERATURE EFFECTS ON pH.
stability study is going on up to 3years for famotidine-USP but the pharmacopeia is revised and tlc test is replaced with hplc test. At the 4th year frequency the product is failed in hplc test .how can i assain the expiry or retest date and can i stop or continue the programme?
Weight balance variation limit during analysis
7 Answers Amneal Pharmaceuticals,
when we use pda detector in hplc instead of uv detector
how do we get usp reference standard potency?
1 Answers Ind Swift Laboratories, MSN Pharma,
why KBr is used in FTIR SPECTROSCOPY and why is it necessary to triturate KBr during FTIR analysis?
25 Answers Hikal, Lupin, Wockhardt,
if out of two set of results of moisture analysis first 1 is passing, second one is failing, is it oos ? or vice a versa first 1 is failing and second one is passing. for information mean value to be reported and in mean value the substance is ppassing, is it oos or normaal mean value to be considered?
Why we are injecting low load standard conc. while analysing Related Substances by HPLC?
What is the formula to determine the concentration of M of a solution given the % transmittance? Use %T = 43.7 as an example.
which chromatographic technique is used for softening hard water?