How to calibrate the magnetic flow meter and how to correct
the flow if it is showing more flow.
Answers were Sorted based on User's Feedback
Answer / maneesh
if the instrument is online calibration can be done by
closing the valve at downstream side.now the meter should
read zero if not recalibrate it.always remember while
calibrating magnetic flow mwter the meter should be filled
with fluid and fluid should be condectieve
Is This Answer Correct ? | 21 Yes | 4 No |
Answer / r.ravikant
There is constant factor for every correspondence flow.
For example,in Krohne Marshall Magnetic flow transmitter
ther is 'GF' factor fix for correspondence flow,so use this
factor.
Is This Answer Correct ? | 17 Yes | 8 No |
Answer / muzaffar ahmed
if the instrument is online calibration can be done by
closing the valve at upstream and downstream side.now the
meter should be
read zero if not recalibrate it.always remember while
calibrating magnetic flow meter the meter should be filled
with fluid and fluid should be conductive.
and other wise condition as such so you should be done by
analog to digital trim.
Is This Answer Correct ? | 3 Yes | 1 No |
Answer / muaz
i ewi
i want to know about supperssion and elevation
Is This Answer Correct ? | 2 Yes | 3 No |
Answer / uc nayak
Magnetic flowmeter calibration method is only by taking known volume of flow through it by taking drop test method.Other than this you can simulate the flow convertor with specific simulator with spefic simulation factor only.
Is This Answer Correct ? | 0 Yes | 1 No |
I am manjunath having exp in maintenance of chemical plants,know i want to do a design course INTOOLS,after doing design course can we get jobs in design field.sinve i dont have design exp..
how to test rotor &stator earth fault in turbine
whats the diff b/w zero pressure and obsolute pressure
How to do ma trim in massflow meter?
In control valve , Failure position : Close If it is Fail close(FC) during failure of AIR.. what will be the initial/ normal position of control valve is it normally open or close? please some one explain soon..
The DCS has the following signal type:
Ple give me "INDUSTRIAL INSTRUMENTATION AND CONTROL S.k. Singh 3th edition" Solution manual to self-Check Quiz and chapter-wise PowerPoint Slides. Home web-site not available... Ple Ple Ple Ple Ple Ple Ple Ple Ple Ple Ple Ple Ple Ple Ple Thanks for all...
how to calculate orifice sizing? and tell me if pipe size bet 50mm t0 200mm then we can use pipe flange??? and above 250mm pipe size should i use D & D/2 Flange? is it right or wrong?
What is PDS and PDMS ?
Draw a simple circuit for a well head control panel used to protect the well against fire, leak, blockage, ESD, & controls the operation of one downhole & one surface safety valve. Explain the operation and controls
which are types of orifice tap ?
How many ohms are required for proper earthing?