how can i calibrate E & h make conductivity transmiter .
unit : mocro siemens
Range :0 to 100 micro siemens
sensor : capacitive type
Answer / karthik
Based on the instrument settings, two known solution shoud
be used for calibration. For example, take 4 mS solution,
immerse the probe and press calibrate option in the display
menu, after 2/3 seconds it will ask like solution1
calibration ok, put solution2?. you have to press yes
button. As per the set(example-7)value, u must put second
known 7 value solution in it. if it is ok. it will finally
ask for calibration over, exit? press yes
Is This Answer Correct ? | 3 Yes | 4 No |
What are all the type of vibration measurement measuring in turbine, what is the output of vibration sensor.
what is maxdna ?. explain about the software details.
why we use the ac supply load cell>?can we use the dc supply load cell?advantages and disadvantage of the ac supply load cell?
What will the new flow factor if D/P Transmitter is RE-RANGED from 25" WC to 50" WC
what should be minimum length and size of impulse line for steam flow orifice and feed water orifice? asked by sanjay parihar
Why we. Don't measure current in parallel? Why we measure current in series and why we measure voltage in parallel only?
2 Answers Almeer Technical Services,
Explain types of valves?
how we can calibrate level transmiiter in field?
How to calibrate Toxic gas detector?
Differentiate between PLC, SCADA and DCS systems
1 Answers Mumias Sugar, SRF, Trimula Industries,
what is RTV function in in process control?
Hi, I'd like to know the purpose of MOS (maintenance override switch). At site, these switches are installed in a stanchion with the transmitter (connected to ESD). thank you.
3 Answers Adani, Adani Wilmar,