how can i calibrate E & h make conductivity transmiter .
unit : mocro siemens
Range :0 to 100 micro siemens
sensor : capacitive type
Answer / karthik
Based on the instrument settings, two known solution shoud
be used for calibration. For example, take 4 mS solution,
immerse the probe and press calibrate option in the display
menu, after 2/3 seconds it will ask like solution1
calibration ok, put solution2?. you have to press yes
button. As per the set(example-7)value, u must put second
known 7 value solution in it. if it is ok. it will finally
ask for calibration over, exit? press yes
| Is This Answer Correct ? | 3 Yes | 4 No |
what is difference between hart signal and analog signal (4 to 20 ma)
how to calibrate TT transmitter
Explain any simple temperature control system.
find the kp ki kd from the transper function G(S)=29.44s+04707/1.5349s2+s
i want all model questions paper for robert bosch
how the 4 to 20 mA signal goes to any controller/indicator,and what is the role of voltage ,12/24 and how to do connection of it, in controller.
where we are used split range control
Define the terms Accuracy, precission, Spam, process tolerance while doing calibration of an instrument? Explaini with an example?
what is linerity and non-linearity?
what is the difference between the microcontroller and PLC?
in plc source and sink meaning?
what is performance test.need defination
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)