Yokogawa mass flow meter shows negative reading...what r the main causes for this problem?...
Answers were Sorted based on User's Feedback
Yokogawa mass flow meter shows negative reading...what r the main causes for this problem?...
Impossible to say, because the technology is not stated, nor are the conditions stated.
Yokogawa is a brand name, not a technology, and the reason for a negative flow indication depends on the technology.
Technologies for mass flow (however inferred) are
- multivariable flow meters based on pressure and temperature compensated differential pressure
- Coriolis flow meters
- Vortex flow meters with temperature and pressure compensation
- Thermal dispersion flow meters
which might or might not be Yokogawa products.
The reasons for a false negative flow indication could be quite different for any of the technologies
Conditions like the expected flow rate vs the observed flow rate, medium, medium temperature and pressure, service history, installation details (vertical or horizontal), flow path, type of pump, process function or contiguous piping and its orientation, and configuration elements like low flow cutoff could contribute to an indicated negative flow rate, or operational method of zeroing.
Because it could be that you actually have reverse flow from a thermal syphon or a pressure syphon.
Carl Ellis
Measure First
Is This Answer Correct ? | 4 Yes | 0 No |
Answer / sarin
HP LP Ports will be interchanged. High pressure is coming in LP Side.
Is This Answer Correct ? | 2 Yes | 2 No |
how does we calibrate a square root flow transmitter ,give an example and explain it step by step like range calculation , calibration hook up series connection , zero check , and how to apply pressure , and if not geting proportional output then what should we do ?
Consider a control valve action,the inlet air pressure is only 6 ksc but physically how it can be operate against 100 ksc line fluid pressure?
Could you explain to me the enraf (servo gauge) mechanism. How it works ? and how it is working on archimedes theory while it's not immersed into the surface? and how can we calibrate it ? Thanx
how to online calibration of remote seal tx.
how can we measure the level measurement with the help of pressure gauge?
can anyone tel the syllabus or previous year question paper for BHAVINI(EIE)2014.
Waht is surge in compressors and how to control it?
how to use level tranasmitter in closed tank application.?
Basically iam Instruemntation Engineer, but Iam interested in Instruments design engineering, so plz guide, is there any course, training or Softwres to become Instrument design engineer??.
0 Answers Honeywell, Tata Steel Limited,
the whole paper was based on Transducers..questions like define repeatability, error,accuracy,precision..den there were many q. on the conversion of Pa to mmHg and other units.. also..der wer few Q. on some level sensors(with diagram)
Reasons for using current as signal
what is modbus?and why use it?what benifit