Why do we use triad cable for RTD to Temp.transmitter connection? Why we don't use 3 core cable?
A triad cable consists of multiple triads, each triad consisting of three conductors(which are twisted together) along with a shied. The fact that the wires are twisted ,and the presence of a shield makes the signal carried by a triad cable less prone to external noise.
A core cable on the other hand consists of individual conductors.The absence of any kind of mechanism to prevent the noise from affecting the signal, makes them unsuitable for carrying any form of analog signal.
If you use a core cable to connect your RTD to the transmitter ,you would get the temperature readings, but it will be susceptible to noise, which is not desirable.
Do let me know if you come across any other reason.
| Is This Answer Correct ? | 22 Yes | 2 No |
I want to use 1/2 inch SS316 tube in my high pressure & temperature (64 kg/cm2 & 500 deg C)after condensate pot in pressure tranmistter. 1) Is it possible to use 1/2" ss316 tube in high pressure and temperature line. Tube manufacturer is local vendor. 2)On sarting the plant, We need to do system fluse that time condesate is not in the condesate pot so steam is directly contact with the tube. We have no exact calculation to know our tube is ok or nor at this time. pleae,reply.
How many loops are there in the power plant? how many closed & open loop in the power plant ? Please give me the exact clarification?
what is the diffrene between DCS System and PLC System
Q-Explain the different load conditions that will produced different offset which may or may not be acceptable to thhe control function
what are the possible reasons if a control valve fails to open/close?
what is flow equation of kg/hr to m3/hr
what is difference between call out and call log
I have energy meter for educational project having 0.5s class but my friend need 0.2s class meter , which one is good for that purpose and what is the actual difference of that two classes , need very easy example also.??
Instrument QA/Qc what is the project quality plan?conduit inspection RFI. which documents attache with package?
why we are using MARSHALLING PANEL? what is the role?
what is XV and KV valves?
what is the difference about HART, and 4-20mA signal?
Civil Engineering (5086)
Mechanical Engineering (4453)
Electrical Engineering (16638)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)