Answer / carl_ellis
Failure to connect with HART can be a laundry list of
issues, the most common of which is the
lack of minimum 230 ohm resistance in the loop.
Other reasons can be
- HART probes connected across the DC power supply, not in
the loop
- failed DC power supply
- electrically noisy DC power supply
- HART modem lacks installed PC driver or internal setup
(serial modems were touchy)
- HART device not at expected HART address (address zero for
point-to-point comm)
- device does not have HART option or HART option has been
disabled.
- HART option replaced by different fieldbus comm card (it's
happened! !)
- Connecting across the test terminals on a pressure
transmitter where there's no HART signal
- HART communicator is dysfunctional or its batteries are
discharged
- lack of DD in communicator/program limits access to the
common commands
Carl Ellis
Measure First
Is This Answer Correct ? | 1 Yes | 0 No |
Dead weight tester & comparator for calibration , which is more accurate , what is the application.
1 Answers Bapco Bahrain Petroleum Company,
types of transformer.. 1ph 3ph or core shell?
Coriolis flow meter coil supply is 24vdc but why still we given 230ac for this transmitter let me explained detailed review please
What are the charcterstics of a P&ID controller block for Governor speed raise or lower for a load sharing functionality between 4 Gas turbines in droop mode.
How Honeywell FSC is functioning and what is its purpose? Can we bypasa it and only use DCS instead of that.
how to calibrate an error for pressure transmitter, make it easy to understand
what is o2 analyzer and how it measured ,what is the role of this analyzer in cfbc boiler...
how can calibrate transmitter online
what is disadvantage of field bus (ff)type of instrument
A temperature sensor (RTD) connected via a temperature transmitter is adjusted to give an output of 4ma at 60 deg and 20 ma at 600 deg. This output (4-20ma) from the transmitter is connected to an IGBT-inverter to control a high speed (-12,000 to +12,000 RPM) DC motor. And in the inverter it is programmed as, for 4ma the motor speed will be -3,000 RPM and for 20ma the motor speed will be +10,000 RPM. The question is : If the motor speed is +5,500 (RPM), what will be the temperature? (-sign indicates that the motion is in reverse direction).
why use 4-20mA in the field of instrumentation?
What is the function of positioner in control valves? does the control valve operate without positioner?