Please explain the difference between instrument range and
calibration range in terms of pressure transmitter?
line design pressure: 93 bar
operating pressure : 82 bar
press Transmitter Range?
calibration range ?
Answers were Sorted based on User's Feedback
Answer / ahmed helmy
- instrument range is the maximum range of instrument.
- Calibration range is at which the instrument calibrated
for example
- if we have a pressure tx instrument range 0:10 bar
- we can calibrate it at any position between 0:10 bar suppose we calibrate it at 0:4 bar its mean its calibration range
is 0:4bar and instrument range 0:10 bar
Based on your case:
- line design pressure=93 bar (max pressure in the line)
- operating pressure=82 bar
so we can install pressure tx range:0 to 120 bar and calibrate the pressure tx in range from 0 to 93 bar
| Is This Answer Correct ? | 19 Yes | 6 No |
Answer / furqan
In addition we also can not calibrate the pressure transmitter as equal to designed line pressure
calibrate the pressure transmitter on 80 percent of designed line pressure.
| Is This Answer Correct ? | 4 Yes | 2 No |
Answer / sri
hi can i know y reasons why cant we choose instrument range as a calibrated range?
| Is This Answer Correct ? | 0 Yes | 1 No |
In s7-200 we use HMI TD-200C , what that C stands for........ plz reply
what are the diff. in servo and regulatory control?
i am working in thermal power plant . if temperature transmitter is not showing reading in dcs then how can we rectify the problem?
If an Instrument has 550 ohm internal resistance and the cable used for that instrument is 0.5 ohm per meter resistance. If the voltage is 12volt.How much length cable can be used?
I am manjunath having exp in maintenance of chemical plants,know i want to do a design course INTOOLS,after doing design course can we get jobs in design field.sinve i dont have design exp..
How to carry out hot loop checking of closed loop pf temperature control valve & temperature transmitter?
A flange is used for instrument connection for a system with a design temperature of 316C (600F) and a design pressure of 1000 kPa (145 psi). What is the class rating of the flange to be used for this application? Explain your answer
IN JB CONSTRUCTION WHY WE ARE USING JUMPPER........CAN U PLEASE TELL ME.....
What is the difference between dry contact and wet contact?
how to calibrate temprature switch and flow switch ?
Any body have know try work calculation ???
How will you test a transistor with a multimeter?
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)