Why we are use 0.005M Sulphuric acid, in control of absorbance
performance test UV Spectrophotometer calibration
Answers were Sorted based on User's Feedback
Answer / manish
Because K2Cr2O7 is only stable in the solution pH below 2.0
| Is This Answer Correct ? | 15 Yes | 0 No |
Answer / ramakrishna
Bcz K2Cr2O7 is easily dissolved in 0.005 M H2SO4 and stable in the solution PH below 2.0
| Is This Answer Correct ? | 12 Yes | 0 No |
Which detector is used in uv-vis spectrophotometer?
Which parameters require to do in tech transfer?
what is RI detector Cut-off Value
how PH an important role in HPLC separation
In the method validation proceedure,lod & loq calculation we use the formulae for LOD=3.3*SD/slope and for LOQ=10*SD/slope,why we use 3.3 & 10 in the formulae?
I want analyze my compound at 325 nm so which standard shall i use to calibrate my HPLC system? How i can assure my hplc detector is suitable or not?
What is difference between NDA & ANDA ?
What is the use of Holmium (III) Oxide???
What is normality
how decide the clining method and cleaning method validation require for this perticular products?
What is the exact reason to use THF and IPA in buffers of Reverse phase HPLC as a solvent?