In the method validation proceedure,lod & loq calculation we use the formulae for LOD=3.3*SD/slope and for LOQ=10*SD/slope,why we use 3.3 & 10 in the formulae?
LOD :- LOD is measured 10 times independent sample of Blank
and Student-t test of 10 times, Degree of freedom 9 (n-1)at
99% confidence label is 3.25 as per Student chart. Also
Ratio of S/N =3:1
LOQ:- LOD is regarded as the lower limit for precise
quantitative Measurement . LOQ defined as concentration
which would a produce a signal 10-12 time. At this point
standard deviation is small enough (Approximately 10-15%) so
that quantified value can be deemed reliable. Also Ratio
S/N= 10:1
Is This Answer Correct ? | 7 Yes | 3 No |
What is the difference between Primay and Secondry Standard ?
6 Answers College School Exams Tests, Neuland,
Difference between the quantitative analysis and qualitative analysis?
Why we have to use KBr pellet technology ?
What is Delay & Dead Volume
IS THERE ANY EQUIPMENT TO CHECK AND CALCULATE THE POLARITY OF A LIQUID?
what are the ideal conditions for coloumn storage
3 Answers Claris Lifesciences, Srikem Laboratories,
Wath is the differenc between two column HPLC wich have the samme support , lenth , pore size, etc.. But different name .for exemple : Spherisorb / lichrosorb
What is the difference between purge septum flow and column flow in gas chromatography?
what is biorelivent dissolution media?
what are the standard methods for HPLC?
why we should take dst factor for below 1%moisture samples
0 Answers Ind Swift Laboratories,
Describe your field experience sample type collected,sample techniques,field measurements taken and equipment used?