how calibrate the hplc detector
Answers were Sorted based on User's Feedback
Answer / m.sivalingeswara rao
Hplc detector calibration to be done to know the acuracy of
the emitted light wave length to the set value as well as
to check the sensitivity of the detecotr to respond with
the concentration of analyte.
To check accuracy generally caffeine solution is used
because it shows maxima and minima at particular wave
lengths(i.e maxima at 206nm,273nm and Minima at 241nm,in
this range we can check accuracy of lamp emitted light).
Sensitivity by permorming linearity test with different
concetration of caffeine solution.
Reproducibility by injecting the replicate injection of
Caffeine solution and measuring the %RSD of the responses
| Is This Answer Correct ? | 19 Yes | 0 No |
Answer / samba siva reddy.p
detector checking with caffiene soln bcz caffiene have 2 max and 1 minima wavelenths,we inject std stock soltn(caffiene soln) with diffrent wavelengths.
201 to 209= 205(maxima)
241 to 249=245(minima)
269 to277=275(maxima)
limits+ or -2
| Is This Answer Correct ? | 2 Yes | 0 No |
Which buffer is used for LC/MS mobile phase & at What Conc.
what is the difference between IR and FTIR instrument analysis
why plane of polarized light rotated by opticaly active compunds?
what is separation techniques.
can i use hplc detector to uplc and why?
pda detector mechanism?
1 Answers DKR, Teva Pharmaceuticals,
what is mean by 21 CFR PART? In that what is 21 ?
why we are analyzing purity on HPLC why not in gc
3 Answers Aurobindo, Reddy Labs,
How to decide restandardisation due or retest or expiry of Volumetric solutions prepared in wet chemistry lab?
effect of pore size, pore volume, partical size, column length, carbon load on retention time? what is carbon load? what is the use?
why phenolpthalin gives pink colour when we titrate aganist sodium hydroxcide & phenolpthalin gives colourless titrate aganist hydrochlric acid which is un burate
Why Methanol used in KF Titration?