how calibrate the hplc detector
Answers were Sorted based on User's Feedback
Answer / m.sivalingeswara rao
Hplc detector calibration to be done to know the acuracy of
the emitted light wave length to the set value as well as
to check the sensitivity of the detecotr to respond with
the concentration of analyte.
To check accuracy generally caffeine solution is used
because it shows maxima and minima at particular wave
lengths(i.e maxima at 206nm,273nm and Minima at 241nm,in
this range we can check accuracy of lamp emitted light).
Sensitivity by permorming linearity test with different
concetration of caffeine solution.
Reproducibility by injecting the replicate injection of
Caffeine solution and measuring the %RSD of the responses
| Is This Answer Correct ? | 19 Yes | 0 No |
Answer / samba siva reddy.p
detector checking with caffiene soln bcz caffiene have 2 max and 1 minima wavelenths,we inject std stock soltn(caffiene soln) with diffrent wavelengths.
201 to 209= 205(maxima)
241 to 249=245(minima)
269 to277=275(maxima)
limits+ or -2
| Is This Answer Correct ? | 2 Yes | 0 No |
What is differences between normality and molarity.in which cases both are used?
why we cannot analysis below 200nm in u.v?
How to establish pKa for a new API molecule ?
2 Answers Neuland, Sun Pharma,
how much wave length of Soduim vapour lamp used in polarimeter
2 Answers Reddy Labs, Torrent Pharma,
What are the factors effect peak shape in HPLC ?
2 Answers College School Exams Tests,
is there any exact definition for Theoritical plates ? chennu srinivas
How require to perform linearity as per ANVISA? What are the acceptance criteria?
if you given one product then which tests you will perform?
how to calculate elemental impurities?
What is delay volume?
How to select Sample concentration in RS analysis?example.
why 0.1M OR 1M Perchloric acid standardisation done after 24 hours?