how calibrate the hplc detector
Answers were Sorted based on User's Feedback
Answer / m.sivalingeswara rao
Hplc detector calibration to be done to know the acuracy of
the emitted light wave length to the set value as well as
to check the sensitivity of the detecotr to respond with
the concentration of analyte.
To check accuracy generally caffeine solution is used
because it shows maxima and minima at particular wave
lengths(i.e maxima at 206nm,273nm and Minima at 241nm,in
this range we can check accuracy of lamp emitted light).
Sensitivity by permorming linearity test with different
concetration of caffeine solution.
Reproducibility by injecting the replicate injection of
Caffeine solution and measuring the %RSD of the responses
Is This Answer Correct ? | 19 Yes | 0 No |
Answer / samba siva reddy.p
detector checking with caffiene soln bcz caffiene have 2 max and 1 minima wavelenths,we inject std stock soltn(caffiene soln) with diffrent wavelengths.
201 to 209= 205(maxima)
241 to 249=245(minima)
269 to277=275(maxima)
limits+ or -2
Is This Answer Correct ? | 2 Yes | 0 No |
Ph meter can show more than 14 ph reading? Why ph range in between 1 to 14 only?
why hydrogen and zero air used for flame ignition in gc?
are there in uv lamb that is chractrestic for wavelenghth 205 nm ?????
1 Answers Beni Suef University, Graviti Pharmaceuticals, Sri Krishna Pharmaceuticals,
Why do we use 500ppb sucrose and 500ppb 1,4 benzoquinone in TOC calibration?
what is the conductivity of hplc grade water?
why we are using k2cr207 in uv calibration ? why we r not using kmno4?
1 Answers Glochem, Sidmak Laboratories,
which solution used for detecting factor of k.f.titroter except water
How do we fix the sample concentaryion in hplc method development. what is the basis?Thanks in advance
4 Answers DRL, Glenmark, Orchid,
in dissolution if one bowl got 70 percent 2nd bowl got 80 percent and 3rd bowl got 90 percent then how proceed?
what is the peak purity concept?
5 Answers Graviti Pharmaceuticals, Vimta Labs,
why we use polystyrene film for ftir calibration?
in each media we require to use SLS? how to proceed?