Answer / m.sivalingeswara rao
Hplc detector calibration to be done to know the acuracy of
the emitted light wave length to the set value as well as
to check the sensitivity of the detecotr to respond with
the concentration of analyte.
To check accuracy generally caffeine solution is used
because it shows maxima and minima at particular wave
lengths(i.e maxima at 206nm,273nm and Minima at 241nm,in
this range we can check accuracy of lamp emitted light).
Sensitivity by permorming linearity test with different
concetration of caffeine solution.
Reproducibility by injecting the replicate injection of
Caffeine solution and measuring the %RSD of the responses
| Is This Answer Correct ? | 19 Yes | 0 No |
Answer / samba siva reddy.p
detector checking with caffiene soln bcz caffiene have 2 max and 1 minima wavelenths,we inject std stock soltn(caffiene soln) with diffrent wavelengths.
201 to 209= 205(maxima)
241 to 249=245(minima)
269 to277=275(maxima)
limits+ or -2
| Is This Answer Correct ? | 2 Yes | 0 No |
What is difference between silica used in TLC and HPLC column
4 Answers Cipla, Meghmani Organics,
Why use only methanol,used determinationof moisture content not use other solvents
what is difference b/w spectroscopy, Spectrometry and Spectrophotometry ?
What is the difference between qualitative and quantitative analysis (hplc)
What is the difference between chromatographic purity and related substances?
defination of wavelength & wave number
how many types oos and oot
which are the guidelines for force degradation studies?
define beer-lambertslaw.
what buffer you can select for acidic nature analyte and how ?
How can you prepare 0.05M Ammonium acetate volumetric solution?
In burette calibration of Karl fischer Instrument, the multiplication factor 1.00693 is using for how much ml of water dispenced. why