How do we fix the sample concentaryion in hplc method
development. what is the basis?Thanks in advance
Answers were Sorted based on User's Feedback
Answer / vipul
For Assay:- Sample concentration should be lower than higher
detector linearity concentration.
For Related substances:- Sample concentration should be
decided on the basis of LOD and LOQ determination
e.g sample concentration 1000ppm
Standard low load concentration 5ppm(For specified imp
limit-0.5%)
LOQ-should be between 0.05% To 0.09% of sample concentration.
| Is This Answer Correct ? | 8 Yes | 2 No |
Answer / pramod
It is also depend upon peak response.
How ever we have to select sample concentration in rs on the basis of LOQ
"Loq should be less than Reporting threshould"
Accordign to that we have to select sample Concentration
| Is This Answer Correct ? | 2 Yes | 0 No |
Answer / rajareddy
based on the peak response ,peak height and LOD and LOQ Signal. Ex:sample concentration 1mg/ml than LOQ value (0.1%mg/ml) LOD value(0.03mg/ml)
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / asha latha
IF ALL IMPURTITIES AT SPEC LEVEL ARE MEETING S/N RATIO MORE
THAN 60 AT THAT PARTICULAR CONCENTRATION AS WEL AS NO
COLUMN OVERLOAD.
| Is This Answer Correct ? | 0 Yes | 6 No |
how to qualify the impurity?
why we use actone for Gpv test in hplc calibration
Which is the highly polar and highly non polar column in HPLC?
Why sulfuric acid is used in sulphated ash / residue on ignition test.
9 Answers CS Pharmaceuticals, Dr Reddys, GVK, HPLC, Micro Labs, Samrudh, Watson Pharmaceuticals, Wings Customer Care,
Why do we use 500ppb sucrose and 500ppb 1,4 benzoquinone in TOC calibration?
how can u seperate polar compounds in hplc ?
1 Answers Meditab, Piramal Healthcare,
How to fix the sample conc. in Related substances test by HPLC?
When we calculate Impurity against impurity std.solution and when we calculate impurity against diluted standard with applying RRF?
What is the procedure for the calibration of ph meter?
What is difference between NDA & ANDA ?
how you start the solubility study?
What is the difference between the optics of UV and PDA Detectors ?