How do we fix the sample concentaryion in hplc method
development. what is the basis?Thanks in advance

Answers were Sorted based on User's Feedback



How do we fix the sample concentaryion in hplc method development. what is the basis?Thanks in adv..

Answer / vipul

For Assay:- Sample concentration should be lower than higher
detector linearity concentration.
For Related substances:- Sample concentration should be
decided on the basis of LOD and LOQ determination
e.g sample concentration 1000ppm
Standard low load concentration 5ppm(For specified imp
limit-0.5%)
LOQ-should be between 0.05% To 0.09% of sample concentration.

Is This Answer Correct ?    8 Yes 2 No

How do we fix the sample concentaryion in hplc method development. what is the basis?Thanks in adv..

Answer / pramod

It is also depend upon peak response.
How ever we have to select sample concentration in rs on the basis of LOQ

"Loq should be less than Reporting threshould"

Accordign to that we have to select sample Concentration

Is This Answer Correct ?    2 Yes 0 No

How do we fix the sample concentaryion in hplc method development. what is the basis?Thanks in adv..

Answer / rajareddy

based on the peak response ,peak height and LOD and LOQ Signal. Ex:sample concentration 1mg/ml than LOQ value (0.1%mg/ml) LOD value(0.03mg/ml)

Is This Answer Correct ?    0 Yes 0 No

How do we fix the sample concentaryion in hplc method development. what is the basis?Thanks in adv..

Answer / asha latha

IF ALL IMPURTITIES AT SPEC LEVEL ARE MEETING S/N RATIO MORE
THAN 60 AT THAT PARTICULAR CONCENTRATION AS WEL AS NO
COLUMN OVERLOAD.

Is This Answer Correct ?    0 Yes 6 No

Post New Answer

More Analytical Chemistry Interview Questions

Why potassium dichromate was using in Uv calibration?

1 Answers  


Which is the highly polar and highly non polar column in HPLC?

0 Answers  


I want analyze my compound at 325 nm so which standard shall i use to calibrate my HPLC system? How i can assure my hplc detector is suitable or not?

3 Answers   AMRI,


why did uyou choose analytical chemistry?

1 Answers  


In Analytical method development related substance method when did follow w/w method when did follow area normalisation method

2 Answers   Dr Reddys, Sun Pharma,






1)What's the meaning of Absorption,give a example. 2)What's the meaning of Adsorption,give a example. 2)what is the difference between Absorption and Adsorption.

0 Answers   VHB Medi Life Sciences,


what is the difference between polarimeter lamp and ir lamp

0 Answers  


In UV calibration we use Holmium perchlorate solution made in 1.4M perchloric acid The question is that 1.4 M perchloric acid we make is it in water or glacial acetic acid?

4 Answers  


what is the conductivity of hplc grade water?

4 Answers  


In case of dissolution test, if there is variation within 6 individual units however the batch is complying as per S1 criteria, is it necessary to do investigation for the same? What should be the criteria for such variation?

2 Answers  


what is primary solution and secondary solution

2 Answers  


why we are using k2cr207 in uv calibration ? why we r not using kmno4?

1 Answers   Glochem, Sidmak Laboratories,


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (302)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1385)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)