Answer / sasikumar.r
1. increase the concentration of organic phase. i.e.,
increase the ratio of organic phase in mobile phase
2. Add little amount of peak modifiying agent TRIETHYLAMINE
3. just REVERSE the column
Is This Answer Correct ? | 15 Yes | 5 No |
Answer / vivek k
Sample solvent stronger than the mobile phase Dissolve
sample in mobile phase or at least reduce the strength of
the sample solvent as much as possible
Sample mass overload Reduce the amount (mass) of sample
injected.
Silanol interaction with amines • Reduce mobile phase pH to
< 3.0
• Increase mobile phase ionic strength.
25mM to 50mM recommended
• Add a competing amine to the mobile phase.
10 mM TEA is usually sufficient.
• Select a stationary phase with a lower silanol activity.
Adsorption of acids on silica • Increase salt
concentration in the mobile phase
25 mM to 50 mM is usually sufficient
• Reduce the pH of the mobile phase to < 3.0
• Add a competing organic acid.
1% acetic acid or 0.1% TFA is usually sufficient
Column void Replace the HPLC column.
Attempts to fill-in the void are seldom worth the effort.
Is This Answer Correct ? | 10 Yes | 0 No |
Answer / shashikumar.v
Add little amount of peak modifiying agent TRIETHYLAMINE
Add 1% acetic acid
Reduce the pH of the mobile phase to < 3.0
Is This Answer Correct ? | 1 Yes | 0 No |
Answer / pragya
By the small addition of organic solvents or by changing the mobile phase ratio with the small change in organic modifiers.
Is This Answer Correct ? | 1 Yes | 0 No |
Answer / anandpcp
Drug is a any chemical or a chemical substance substance
which are used for the specific purpose cure, prevention
etc...i.e, treatment of any disease or disorder for humen or
animal.
Is This Answer Correct ? | 1 Yes | 5 No |
Answer / dhanalakshmi
You can use Peak height or Peak area to calculate the
concentration and the results are no big different. My
experience is to see which one is sharp on the chromatogram.
E.g. if a very sharp peak was found then I will use the peak
height, if a flat peak then I will use the peak area for
calculation
Is This Answer Correct ? | 3 Yes | 15 No |
which procedure we follow if 6th month of accelerated sample fail?
various methos to calculate precision in validation methods in HPLC?
What is the difference between chiral cel OD-H & OD-RH ?
why ph electrode is storing in KCL solution?Why the pH max is 14.0?
Difference Between End Cap HPLC Column and Non End Cap Column
what is the structure of DOWEX-50
What is rs test why we are performing rs test?
what is the diff betwen calibration and system suitability?
How can you calculate manually "Slope" in pH calibration?
My question about gas chromatography sulfur chemiluminsecence detector. I test unknown sample gas by GC-SCD (calibrated ) and the result of *H2S is 279 PPM , *but when I test the same sample with the GC-TCD (calibrated ) the value of *H2S is 2500 PPM . I'd like to inform you that both GCs are calibrated and have very good operation conditions with stable parameters . the question is if the sample gas with higher H2S over detection limits of SCD detector (1000 ppm). why the result it 279 ppm Best regards
what is the importance of peak purity in HPLC and how we can calculate through manul(not software)?
In Agilent HPLC calibration using Empower 3 Software, What is the criteria for GPV Baseline starting point? i.e., Either it can start from -0.00/+0.00 or any range is there like -0.05 to +0.05 or ......etc?