control of resolution in calibration of IR
Answers were Sorted based on User's Feedback
Answer / nirav intas pharma
control of resolution in IR:
Take spectrum of polystyrine film,
the abs minima at 2870 cm- & abs maxima at 2849.5cm- is
greater than 0.33
the abs minima at 1589 cm- & abs maxima at 1583 cm- is
greater than 0.08
| Is This Answer Correct ? | 12 Yes | 0 No |
Answer / kuchela rao athota
MAXIMUM ABSORPTION AT ABPUT 1583CM-1 SHOULD BE GREATER THAN
12%
MAXIMUM ABSORPTION AT ABPUT 2850CM-1 SHOULD BE GREATER THAN
18%
| Is This Answer Correct ? | 5 Yes | 0 No |
Answer / arpit vashi
the difference between two percentage transmittance
(2870 cm-1 and 2849.5 cm-1) must be grater than 18%
the difference between two percentage transmittance
(1589 cm-1 and 1583 cm-1) must be grater than 10%
| Is This Answer Correct ? | 1 Yes | 0 No |
what is the disso medium for tropsiun chloride
why in GC calibration of linearity regression coefficient should not be less than 0.999
what are the different methods of analysing related substances test,what are LOD AND LOQ .HOW TO REPORT IMPURITIES WHEN LOD AND LOQ LIMITS R SPECIFIED
What is the principle of HPLC, GC, LCMS, GCMS,LC QQQ, GC QQQ, LC Q TOF and GC Q TOF. What are the applications and Specificity?
why require to add enzymes in capsules dissolution only? is it require to tablet also?
why we are using benzene, anyline in acetic anhdride assay titration method?
Please tell me about the pH of Polycaboxylic ether is it in the 5-6 range ever or more than 6
0 Answers Apex Encon Projects Private Limited, Cipla,
why 1.2%kcl solution is using in limit os straylight?and why this test used for uv calibration?
1 Answers Graviti Pharmaceuticals, Reddy Labs, Vimta Labs,
Why linearity of detector performed in HPLC calibration
What is UV cut off?
what is the diffrence between absorbance and absorption in the beer,s lamber formula ?
How precise should be the linearity in HPLC method validation?