How to calibrate a orifice differential pressure flow
transmitter in process?
Answer / ayaz
first communicate ccr tag conform and turn MOS if required
when 5 way manifold
1 hp side close
2 equalizer open
3 lp side close
and then hp and lp drain or vent should be open
the transmitter will show zero if not calibrate it
first adjust the zero
2nd second span check and adjust
and then service
1 drain or vent close
2 equalizer close
3 and then hp side open and lp side open
| Is This Answer Correct ? | 3 Yes | 0 No |
How to convert milliampares into flow??
Why do we floating DC in DC system.
why k factor is important to know in flow transmitter. if we don,t know k factor what will be issue we can face .
1 Answers Flour Tech Engineers (P) Ltd,
why twisted pair of wires is used in instrumentation ?
Can all intrinsic safe transmitters are provided with inbuilt zener barriers? why
Explain the working of an enraf level gauge.
please tell me how to convert double acting ytc positioner yt-1200 into single acting ytc positioner ?
what is droop mode n frequency mode in wood ward governor?
give me all the details about the interview and which type of question being asked
What is SLPC & MLPC? How it works?
What is Loop Test & What is Loop Checking? Please i need answer for these questions.
WHAT IS THE DIFFERENCE BETWEEN DCS & PLC.
Civil Engineering (5086)
Mechanical Engineering (4453)
Electrical Engineering (16638)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)