Answer / abdulkalim
Parameters involved during initial stage of Analytical
Method development:
1. Optimization of Mobile Phase
2. Selection of Buffers
3. Selection of Ion pair reagent
4. pH of buffer/mobile phase
5. Composition of Mobile Phase
6. Selection of Column
7. Optimization of Analyte signal
8. Objectives of Seperation
8.1 Analysis time
8.2 Resolution
8.3 k factor (Capacity factor)
8.4 Peak height
8.5 Asyymetry
8.6 Theoritical Plates
9. Flow rate
| Is This Answer Correct ? | 90 Yes | 7 No |
Answer / ramana
Parameters involved during initial stage of Analytical
Method development:
1. Solubility
2. Pka Value
3. Selection of Buffers
4. pH of buffer/mobile phase
5. Composition of Mobile Phase
6. Selection of Column
7. Optimization of Analyte signal
8. Objectives of Seperation
8.1 Analysis time
8.2 Resolution
8.3 k factor (Capacity factor)
8.4 Peak height
8.5 Asyymetry
8.6 Theoritical Plates
9. Flow rate
| Is This Answer Correct ? | 54 Yes | 2 No |
Answer / yojana
1. Check solubility,if sample dissolved in water then use
reverse phase,othervise use normal phase.
2.wavelength selection
3.selection of mobile phase,buffer.
4. pH OF BUFFER &MOBILE PHASE
5. selection of column
6. check retention time
7 .Flow rate
8. check system suitability parameter like
i)theoritical plates
ii) Assymetry
iii) Resolution
iv) capacitor fator
| Is This Answer Correct ? | 31 Yes | 2 No |
Answer / vinod kumar, nektar therapeuti
1.If molecular weight less than 2000. based on their
solubility follow as below
2.If water soluble and non ionic follow reverse phase.
3.If water soluble and ionic follow reverse phase ion pair
method.
4.If organic soluble follow normal phase.
5.If molecular weight is more than 2000.Based on their
solubility follow as below.
6.If organic soluble follow gel permeation chromatography
7.If water soluble ion exchange or gel filtration.
8.Once the technique is fixed optimise column.
9.Optimise buffers
10.optimise pH of buffers
11.optimise concentration of buffers.
12.optimise type of organic solvent.
13.optimise column temperature
14.optimise gradient programme to improve resolution and to
elute out late eluting peaks.
| Is This Answer Correct ? | 19 Yes | 11 No |
Answer / sivannarayana
1.Solubility
2.Find out the UV scan For wave length
3.optimaize Buffer & pH Value Optimaize
4.Optimaize Column Temparature & optimaize sample cooler Temparature.
5.optimaize the GP For better Resolution
| Is This Answer Correct ? | 4 Yes | 0 No |
Answer / anand patel
as per my knowledge system suitability test is the method
delopment parameter
| Is This Answer Correct ? | 10 Yes | 19 No |
differnc btw c18 and c-8 column
What is the thickness of polystyrene film?
8 Answers Pharmaffiliates Analytics and Synthetics,
In partical size by malven,which parameter is generlly change for development of dry and wet method.
Describe 5 components of SFC ?
1 Answers College School Exams Tests,
what is the meaning of stress study?
I got 88.0 % assay for a drug in HPLC analysis for same injection i got 40.0% related substances is it possible? what is the relation between results of assay and purity?
in hplc calibration injector linearity pass,it means inj and det is good. why to perform detector linearity is it required, if it is required,what is difference between injector and detector linearity in hplc calibration
Why perform dissolution multimedia study in dissolution Development
how you confirm the assay method?
How to Calibrate NMR instrument ?
Is it compulsory to prepare KBr disc with sample in identification test by FTIR?
IF WE INCREASE FLOW RATE RT DECRESE.WHY