Answer / abdulkalim
Parameters involved during initial stage of Analytical
Method development:
1. Optimization of Mobile Phase
2. Selection of Buffers
3. Selection of Ion pair reagent
4. pH of buffer/mobile phase
5. Composition of Mobile Phase
6. Selection of Column
7. Optimization of Analyte signal
8. Objectives of Seperation
8.1 Analysis time
8.2 Resolution
8.3 k factor (Capacity factor)
8.4 Peak height
8.5 Asyymetry
8.6 Theoritical Plates
9. Flow rate
| Is This Answer Correct ? | 90 Yes | 7 No |
Answer / ramana
Parameters involved during initial stage of Analytical
Method development:
1. Solubility
2. Pka Value
3. Selection of Buffers
4. pH of buffer/mobile phase
5. Composition of Mobile Phase
6. Selection of Column
7. Optimization of Analyte signal
8. Objectives of Seperation
8.1 Analysis time
8.2 Resolution
8.3 k factor (Capacity factor)
8.4 Peak height
8.5 Asyymetry
8.6 Theoritical Plates
9. Flow rate
| Is This Answer Correct ? | 54 Yes | 2 No |
Answer / yojana
1. Check solubility,if sample dissolved in water then use
reverse phase,othervise use normal phase.
2.wavelength selection
3.selection of mobile phase,buffer.
4. pH OF BUFFER &MOBILE PHASE
5. selection of column
6. check retention time
7 .Flow rate
8. check system suitability parameter like
i)theoritical plates
ii) Assymetry
iii) Resolution
iv) capacitor fator
| Is This Answer Correct ? | 31 Yes | 2 No |
Answer / vinod kumar, nektar therapeuti
1.If molecular weight less than 2000. based on their
solubility follow as below
2.If water soluble and non ionic follow reverse phase.
3.If water soluble and ionic follow reverse phase ion pair
method.
4.If organic soluble follow normal phase.
5.If molecular weight is more than 2000.Based on their
solubility follow as below.
6.If organic soluble follow gel permeation chromatography
7.If water soluble ion exchange or gel filtration.
8.Once the technique is fixed optimise column.
9.Optimise buffers
10.optimise pH of buffers
11.optimise concentration of buffers.
12.optimise type of organic solvent.
13.optimise column temperature
14.optimise gradient programme to improve resolution and to
elute out late eluting peaks.
| Is This Answer Correct ? | 19 Yes | 11 No |
Answer / sivannarayana
1.Solubility
2.Find out the UV scan For wave length
3.optimaize Buffer & pH Value Optimaize
4.Optimaize Column Temparature & optimaize sample cooler Temparature.
5.optimaize the GP For better Resolution
| Is This Answer Correct ? | 4 Yes | 0 No |
Answer / anand patel
as per my knowledge system suitability test is the method
delopment parameter
| Is This Answer Correct ? | 10 Yes | 19 No |
What is the difference between known and unknown impurities?
From peak purity spectrum, where or at which point peak purity angle & purity threshold is checked?
What is the difference between void volume and delay volume in HPLC?
What is the role of SO2,I2, Methanolic hydroxide and any base in Karlficher titration
2 Answers Medley Pharmaceuticals,
hii, i'm had a sample of 1L that gone through extraction & evaporation. i used 0.5ml of methanol for reconstitution before injecting the sample to HPLC. i made a calibration curve using a standard solution but now i'm confused about how to display the concentration i got from the sample. for ex. i got 0.2ug/ml how i'm gonna explane this is for the 1L ? please help .. i'll really appreciate it thanks
what is a placebo peak
2 Answers GMP Pharma, Optimus,
What is boiling point ? & Explain.
what is lod and loq ?,why use k2cr2o7 , kcl h2so4 in uv calibration ?,why use benzophenone & caffene acetone in hplc calibration ?,what is leading peak in hplc ?why we do the calibration of limit of stry light in hplc & uv ?
What are the tests used in Hplc?Name any two.
How to select Sample concentration in RS analysis?example.
why we are using k2cr207 in uv calibration ? why we r not using kmno4?
1 Answers Glochem, Sidmak Laboratories,
Which detector is used in uv-vis spectrophotometer?