Answer / k.praveen kumar macleods pharm
If the method is linear or not from certain range suppose
in assay validation the range is 50% to 150% we are
injecting minimum 6 levels and also Calculate the regration
line the correlation coefficient should be less than 0.999
Here regration line equation is y=mx+c
where y = Regration line
m = Slope
c = Intercept.
and also calculate residual sum of squares from theorical
and practical
| Is This Answer Correct ? | 13 Yes | 4 No |
Linearity is done in hplc method validation to show that the
method is linear through out the concentration range i.e.
the hplc method can give results according to the set
concentration , for instance if 1x is the area obtained for
50% solution then for 100% solution it should bee 2X i.e.
and so on. So for every hplc method lineariy is a must
| Is This Answer Correct ? | 6 Yes | 0 No |
how to prepare low ppm solution from high ppm solution ?
what is the limit of Correlation for IR test compare to the standard and test
why we get negative values in uv when i placed sample which has some concetration but not in standard
is there any exact definition for Theoritical plates ? chennu srinivas
In Analytical method development related substance method when did follow w/w method when did follow area normalisation method
diference b/w packed &capilary columns
in HPLC injection volume is increase ,R.T will increase ?
5 Answers Claris Lifesciences,
what are risk assement in the analytical qbd?
why require to add enzymes in capsules dissolution only? is it require to tablet also?
Why dissolution test performed on six jar?
In lcms analysis some times we get sodium and potassium adducts. if the compound is havine sodium ions how can we confirm wether sodium is from sample or it is a adduct?
2 Answers Teva Pharmaceuticals,
if out of two set of results of moisture analysis first 1 is passing, second one is failing, is it oos ? or vice a versa first 1 is failing and second one is passing. for information mean value to be reported and in mean value the substance is ppassing, is it oos or normaal mean value to be considered?