why we wre doing linearity siudiec in hplc method validation
Answers were Sorted based on User's Feedback
Answer / k.praveen kumar macleods pharm
If the method is linear or not from certain range suppose
in assay validation the range is 50% to 150% we are
injecting minimum 6 levels and also Calculate the regration
line the correlation coefficient should be less than 0.999
Here regration line equation is y=mx+c
where y = Regration line
m = Slope
c = Intercept.
and also calculate residual sum of squares from theorical
and practical
| Is This Answer Correct ? | 13 Yes | 4 No |
Linearity is done in hplc method validation to show that the
method is linear through out the concentration range i.e.
the hplc method can give results according to the set
concentration , for instance if 1x is the area obtained for
50% solution then for 100% solution it should bee 2X i.e.
and so on. So for every hplc method lineariy is a must
| Is This Answer Correct ? | 6 Yes | 0 No |
why RID detector is used for selective materials like Mannitol, Hydroxyethyl Starch and Sorbitol?
If i performing one analysis on C8column then the same analysis done on C18 column then what will happen about Rt;Resulution;Peak shape Explain ?
why don't we call POH meter? As we all as PH meter
why we are using hexane in calibration of number of drop per mL
What is the column performance and column efficiency?
How to prepare 10% HCl from 36.5% HCl solution.
difference between HPLC & LC-MS
could negative ions be produced by bombardment process in mass spectrometry?
IN pH CALIBRATION,CALIBRATION IS DONE FOR pH METER OR ELECTRODE?HOW WILL YOU JUDGE THAT THE METER TO BE CALIBRATED WITH 4,7,9.2 BUFFERS?EXPLAIN TEMPERATURE EFFECTS ON pH.
How to calculate coreletion coefficient
why 0.1M OR 1M Perchloric acid standardisation done after 24 hours?
what is difference between assayand potency