What is calibration procedure of gredient HPLC pump
Answers were Sorted based on User's Feedback
Answer / abhijeet gavali
You have two different methods for it
1.0 Use Propyl parabean in methanol and pure methanol
2.0 use 0.05% solution of acetone in water and water
| Is This Answer Correct ? | 14 Yes | 3 No |
Mobile phase A is 5.6mg/l of propylparaben in methanol (port
A and B)
Mobile phase B is 100% methanol (port C and D)
Run the test for 20 min by using curve 11 method
sample is Methanol
Acceptance criteria:
A/B/C is between 9.5 % to 10.5
A/B/D is between 9.5 % to 10.5
| Is This Answer Correct ? | 13 Yes | 5 No |
Answer / kuchela rao athota
in waters gradient program
TIME PUMPA% pUMPB% PUMPC% PUMPD% CURVE
0.00 50% 50% 0 0
2.00 0% 0% 50 50 11
6.00 50% 50% 0% 0 11
10.00 45% 45% 10 0 11
12.00 50% 50% 0 0% 11
14.00 45% 45% 0 10 11
16.00 50% 50% 0 0 11
18.00 50% 50% 0 0 11
| Is This Answer Correct ? | 8 Yes | 6 No |
Answer / prerna
Gradient Programme for port A and C :
00 : 100
00 : 100
25 : 75
25 : 75
50 : 50
50 : 50
75 : 25
75 : 25
100 : 00
100 : 00
Above programme also run for port B and D
An increasing stepwise piramid will appear. Integrate that
peak and also split each step of the peak.
Find out its height and area of height
By using the formula ( which I don't remeber ), mean flow
of solvent at each step.
Flow : Limit
25 (24-26)
50 (49-51)
75 (74-76)
100 (100)
If the observed value is within the acceptance limit then
that pump passes the test.
| Is This Answer Correct ? | 10 Yes | 10 No |
why we conclude the area% for the chromatographic purity?
1 Answers Amneal Pharmaceuticals, Lupin,
what is the principle involved in turbidometry and nephelometry ?
0 Answers Graviti Pharmaceuticals,
difference between IR & NMR spectroscopy ?
Why karl fisher factor is 5 ?
how to set sample and standard concentration in RS method?
Why KBr is used in FTIR ?
process of Diclofenac sodium,IP.
Why the same type of columns showing different RT for a single product?
How can you calculate manually "Slope" in pH calibration?
why 2Theta value is consider in the XRPD spectra and how to consider the 2 Theta value
why disodium tartarate is used for autotitrator calibration
why using factor 1.00639 in burette calibration of kf