what is the defination of calibration & verification?
Answers were Sorted based on User's Feedback
Answer / abdulkalim
Definitions:
------------
Calibration: This is an act of checking or adjusting by
comparing with known standard for accuacy of the instrument
to meet the defined criteria with a specificed degree of
confidence.
Performance Verification:
-------------------------
It is an activity to document that the instrument
consistently performs accordingly to the predefined
specifications which is appropriate for intended use.
Differences:
------------
Calibration is the process where we can do adjustment is a
part of calibration but not necessary.
Where as Peformance Verification is only the act to check
the Performance of the instrument.
Example
-------
Using Salicylic acid and Prednisone Tablets for Dissolution
Apparatus (Chemical testing) is a Performance Verification
test rather being called as "Calibration" since we are
checking the performance of the equipment whereas we are
not doining any adjustment at any point of time .Hence it
is eventually called as "Performance Verification"
| Is This Answer Correct ? | 34 Yes | 1 No |
Answer / prasad
calibration is to check that perticular instrument is
working under specified limits and conditions
| Is This Answer Correct ? | 37 Yes | 9 No |
Answer / gayathri
Calibration is confirming periodically whether the
instrument is working properly with respect to the
particular parameters and ensuring that the same can be
used in the future.
| Is This Answer Correct ? | 19 Yes | 7 No |
Answer / alpesh patel
Calibration is to verify the particular instrument with the
comparition of known standard or instrument.
| Is This Answer Correct ? | 13 Yes | 5 No |
Answer / deepakm
calibration is the set of operations under specified
conditions that establishes a relationship between the
values of quantities indicated by the measuring instrument
or system and the corresponding values indicate by the
standard and to adjust if any deviations are observed to
bring into tolerance limitsi.e. this is basically the
comparision of process against a standard.
verification means to verify something against some set of
standards to find out whether it meets the requirements or
not.it is only an act of checking bt no adjustment has to be
done.
| Is This Answer Correct ? | 5 Yes | 0 No |
Answer / pankaj pawar
Calibration is simply the process of comparing the unknown with a reference standard and reporting the results. For example:
Applied= 1.30V, Indicated= 1.26V (or Error= -0.04V)
Calibration may include adjustment to correct any deviation from the value of the standard.
Verification, as it relates to calibration, is the comparison of the results against a specification, usually the manufacturer's published performance figures for the product. (e.g. Error= -0.04V, Spec= ±0.03V, "FAIL"). Some cal labs include a spec status statement on their Certificate of Calibration. (i.e. the item did/did not comply with a particular spec).
| Is This Answer Correct ? | 5 Yes | 0 No |
Answer / basha
calibraion is the checking the instrument with reference
standards they are in limit or out of limit
| Is This Answer Correct ? | 5 Yes | 2 No |
Answer / balakrishna
Calibration: it is a process where an instrument is set to
provide a correct response to a given set of conditions.
verification:it is a process for such instruments is one of
challenging the performance of the instruments by measuring
known standards.
| Is This Answer Correct ? | 3 Yes | 7 No |
In lcms analysis some times we get sodium and potassium adducts. if the compound is havine sodium ions how can we confirm wether sodium is from sample or it is a adduct?
2 Answers Teva Pharmaceuticals,
What is NPOC method in TOC?
What is delay volume?
why phenolpthalin gives pink colour when we titrate aganist sodium hydroxcide & phenolpthalin gives colourless titrate aganist hydrochlric acid which is un burate
what is mean by extactable and leachable study?
WHAT IS UNIT OF ABSORBANCE AREA IN HPLC OR UV DETECTER
What is void volume & dead volueme
i set electrode to 7 buffer than i used it to check 4 buffer and 10 buffer if the results are not as expected than is it better to set the buffer again to 4 buffer or is there any other method available please let me know ...
what is the limit of Correlation for IR test compare to the standard and test
what is mass spectra ? why it is necessary ?
why we use glass fiber filters use in some situation?
why do u get different drug release while doing calibration of dissolution test appratus 1& 2(paddle & basket)?
1 Answers HDFC, Medrich, Twilight Litaka Pharma,