Answer / murali reddy
We have to 're check why it has failed.For example wavelength accuracy failed means we have to do impact analysis
That is before calibration of HPLC ..Which products are analysed
The same products (same test) had to repeat on another system
Like this last 5 samples need to analysed.if results are ok
We can say no impact.Then after knowing actual problem
Re calibration required n proper justification should be given.
| Is This Answer Correct ? | 6 Yes | 0 No |
Answer / pranav
IN CASE OF OUT OF CALIBRATION AN EXTERNAL AGENCY WILL
CALIBRATE THE INSTRUMENT
| Is This Answer Correct ? | 2 Yes | 4 No |
What is Rf value& Rx value in TLC?
what is the Bracketing & Matrixing.
How to make up 100ml of a solution cintaining compound A at 95µg/ml and compound B at30µg/ml for an HPLC reference standard.
WHAT IS THE DIFFERENT BETWEEN PURITY, POTENCY AND ASSAY
why we use sodium salicylate in uv light calibration at TLC chamber
1 Answers Aurobindo, Lee Pharma,
What is difference between FID and TCD in gas chromatography
why we are using k2cr207 in uv calibration ? why we r not using kmno4?
1 Answers Glochem, Sidmak Laboratories,
Why potassium hydrogen pthalate is used for perchloric acid And naoh. Why it is used for both acid and base
6 Answers Vijayasri Chemicals,
Why dissolution test apparatus calibration with salycilic acid tablets was stopped by FDA&USP from the year of 2009?
In Assy by potentio metry titration method how select MET U ,DET U, SET U,
In C8,C18....etc have 4&5 etc... microns Columns is there like that how to separate Chiral columns as per Micron size?
Why we are use 0.005M Sulphuric acid, in control of absorbance performance test UV Spectrophotometer calibration