How to calculate LOD & LOQ in HPLC Validation Method

Answers were Sorted based on User's Feedback



How to calculate LOD & LOQ in HPLC Validation Method..

Answer / abdulkalim

Based on standard deviation of the response and the slope
obtained will be calculated as:

Limit of Detection (LOD):
DL= 3(SD/Slope)


Limit of quantification(LOQ):
DL=10(SD/Slope)

Is This Answer Correct ?    231 Yes 20 No

How to calculate LOD & LOQ in HPLC Validation Method..

Answer / m

There are three different method indiated in ICH guidelines
for determination of LOD and LOQ.

1 Visual evaution method - For non instrument and sometimes
for instrumental also.

2 Based on residual standard deviation of the response and
the slope
obtained will be calculated as:

Limit of Detection (LOD):
DL= 3(SD/Slope)


Limit of quantification(LOQ):
DL=10(SD/Slope)

3 Signal to noise ratio method

But residual standard deviation method is more reliable
method for determination of LOD and LOQ value

Is This Answer Correct ?    138 Yes 17 No

How to calculate LOD & LOQ in HPLC Validation Method..

Answer / balaji

There are three different method indiated in ICH guidelines
for determination of LOD and LOQ.

1 Visual evaution method - For non instrument and sometimes
for instrumental also.

2 Slope Method:Based on residual standard deviation of the
response and
the slope
obtained will be calculated as:

Limit of Detection (LOD):
DL= 3(RESIDUAL SD/Slope)


Limit of quantification(LOQ):
DL=10(RESIDUAL SD/Slope)

3 Signal to noise ratio method
Limit of Detection (LOD):
S/N NLT 3

Limit of quantification(LOQ):
S/N NLT 10
Slope and S/N methods are more reliable
method for determination of LOD and LOQ value

Is This Answer Correct ?    58 Yes 11 No

How to calculate LOD & LOQ in HPLC Validation Method..

Answer / sunil.madhukar rao kulkarni

We can calculate LOD and LOQ values using linearity
graph,insted of sigma value STEYX can be considered for
more acuurate vale of LOD and LOQ

where STEYX : Standard error of the pridected Y-value for
each X in a regression

Final calulation given as below :

LOD = STEYX x 3/ Slope

LOQ = STEYX x 10 /Slope

Is This Answer Correct ?    44 Yes 10 No

How to calculate LOD & LOQ in HPLC Validation Method..

Answer / ashutosh kumar srivastava

Signal of noise ratio X 3 = LOD
LOD X 3.33 =LOQ

Is This Answer Correct ?    33 Yes 12 No

How to calculate LOD & LOQ in HPLC Validation Method..

Answer / bhupalreddy

Limit of Detection (LOD):
• Based on visual limitation
– Visual evaluation may be used for non-instrumental
methods but may also be used with instrumental methods.
– The detection limit is determined by the analysis
of samples with known concentrations of analyte and by
establishing the minimum level at which the analyte can be
reliably detected.
• Based on signal-to-noise
– This approach can only be applied to analytical
procedures which exhibit baseline noise.
– Determination of the signal-to-noise ratios is
performed by comparing measured signals from samples with
known low concentrations of analyte with those of blank
samples an establishing the minimum concentration at which
the analyte can be reliably detected. A signal-to-noise
ratio between 3 or 2:1 is generally considered acceptable
for estimating the detection limit.
• Based on the standard deviation of the response
and the slope
– The detection limit = 3.3 s / S
– Where s is the standard deviation of the response
and S is slope of the calibration curve.
– The estimate of S may be carried out in a variety
of ways, for example:
• Based on the standard deviation of the blank
• Based on the calibration curve


Limit of Quantification (LOQ):
• Based on visual limitation
– Visual evaluation may be used for non-instrumental
methods but may also be used with instrumental methods.
– The quantification limit is determined by the
analysis of samples with known concentrations of analyte
and by establishing the minimum level at which the analyte
can be quantified with acceptable accuracy and precision.
• Based on signal-to-noise
– This approach can only be applied to analytical
procedures which exhibit baseline noise.
– Determination of the signal-to-noise ratios is
performed by comparing measured signals from samples with
known low concentrations of analyte with those of blank
samples an establishing the minimum concentration at which
the analyte can be reliably quantified. A typical signal-to-
noise ratios is 10:1.
• Based on the standard deviation of the response
and the slope
– The quantification limit = 10s / S
– Where s is the standard deviation of the response
and S is slope of the calibration curve.
– The estimate of S may be carried out in a variety
of ways, for example:
• Based on the standard deviation of the blank
• Based on the calibration curve

Is This Answer Correct ?    21 Yes 4 No

How to calculate LOD & LOQ in HPLC Validation Method..

Answer / dr.praveen choubey

by HPLC LOD @ LOQ calculated by S/N ratio that is the best
method.LOD s/N NLT 3 and LOQ S/N NLT 10 and also calculated
by visual evalution.

Is This Answer Correct ?    11 Yes 2 No

How to calculate LOD & LOQ in HPLC Validation Method..

Answer / nikhil gandhi

Based on ICH Guideline there are many method given to
identify LOD and LOQ, but most useful method is slope and
residual standard deviation is used.

LOD : 3.3 X sigma/ Slope
LOQ : 10 X sigma/ Slope

Sigma : Residual Standard Deviation

Is This Answer Correct ?    12 Yes 4 No

How to calculate LOD & LOQ in HPLC Validation Method..

Answer / rakesh p. kotkar

Based on standard deviation of the response and the slope
obtained will be calculated as:

Limit of Detection (LOD):
DL= 3.3(ASD/Slope)


Limit of quantification(LOQ):
DL=10(ASD/Slope)

Is This Answer Correct ?    10 Yes 3 No

How to calculate LOD & LOQ in HPLC Validation Method..

Answer / pruthvi

why we have to take 3 for LOD and 10 for LOQ? why not other numbers?

Is This Answer Correct ?    6 Yes 0 No

Post New Answer

More Analytical Chemistry Interview Questions

why sucrose not used in polarimeter calibration?

2 Answers   Cipla,


Hi sir if any product monograph not given known impurities (may be 5 impurities) specifications then how we require to proceed for inhouse formulation?

0 Answers  


What is the limit of minimum area in in hplc RS anlysis

1 Answers   Lupin,


In partical size by malven,which parameter is generlly change for development of dry and wet method.

1 Answers  


Why we use methanol as solvent in kf standardisation. Why not other solvent?

2 Answers   Caplin Point, Epione Labs, Hetero,






Why acetonitrile and water are used as extraction solvent when analysis melamine? I thought they are miscible and won't be able to separate...

0 Answers  


what is specificity?

4 Answers   Amneal Pharmaceuticals, Aurobindo,


What is split ratio in Gc? Splitless? how requirr to select?

0 Answers  


India is. Under which zone as per QF1.?

4 Answers   Acebright, Aurobindo, Gland Pharma,


How are PDA Detector work in HPLC

7 Answers   Agilent,


why we calibrate ph meter and balance daily.....but hplc and uv requires quaterly?

5 Answers   Apotex, UPL,


Explain what is a base line?

1 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)