How to calculate LOD & LOQ in HPLC Validation Method
Answers were Sorted based on User's Feedback
Answer / abdulkalim
Based on standard deviation of the response and the slope
obtained will be calculated as:
Limit of Detection (LOD):
DL= 3(SD/Slope)
Limit of quantification(LOQ):
DL=10(SD/Slope)
| Is This Answer Correct ? | 231 Yes | 20 No |
Answer / m
There are three different method indiated in ICH guidelines
for determination of LOD and LOQ.
1 Visual evaution method - For non instrument and sometimes
for instrumental also.
2 Based on residual standard deviation of the response and
the slope
obtained will be calculated as:
Limit of Detection (LOD):
DL= 3(SD/Slope)
Limit of quantification(LOQ):
DL=10(SD/Slope)
3 Signal to noise ratio method
But residual standard deviation method is more reliable
method for determination of LOD and LOQ value
| Is This Answer Correct ? | 138 Yes | 17 No |
There are three different method indiated in ICH guidelines
for determination of LOD and LOQ.
1 Visual evaution method - For non instrument and sometimes
for instrumental also.
2 Slope Method:Based on residual standard deviation of the
response and
the slope
obtained will be calculated as:
Limit of Detection (LOD):
DL= 3(RESIDUAL SD/Slope)
Limit of quantification(LOQ):
DL=10(RESIDUAL SD/Slope)
3 Signal to noise ratio method
Limit of Detection (LOD):
S/N NLT 3
Limit of quantification(LOQ):
S/N NLT 10
Slope and S/N methods are more reliable
method for determination of LOD and LOQ value
| Is This Answer Correct ? | 58 Yes | 11 No |
Answer / sunil.madhukar rao kulkarni
We can calculate LOD and LOQ values using linearity
graph,insted of sigma value STEYX can be considered for
more acuurate vale of LOD and LOQ
where STEYX : Standard error of the pridected Y-value for
each X in a regression
Final calulation given as below :
LOD = STEYX x 3/ Slope
LOQ = STEYX x 10 /Slope
| Is This Answer Correct ? | 44 Yes | 10 No |
Answer / ashutosh kumar srivastava
Signal of noise ratio X 3 = LOD
LOD X 3.33 =LOQ
| Is This Answer Correct ? | 33 Yes | 12 No |
Limit of Detection (LOD):
• Based on visual limitation
– Visual evaluation may be used for non-instrumental
methods but may also be used with instrumental methods.
– The detection limit is determined by the analysis
of samples with known concentrations of analyte and by
establishing the minimum level at which the analyte can be
reliably detected.
• Based on signal-to-noise
– This approach can only be applied to analytical
procedures which exhibit baseline noise.
– Determination of the signal-to-noise ratios is
performed by comparing measured signals from samples with
known low concentrations of analyte with those of blank
samples an establishing the minimum concentration at which
the analyte can be reliably detected. A signal-to-noise
ratio between 3 or 2:1 is generally considered acceptable
for estimating the detection limit.
• Based on the standard deviation of the response
and the slope
– The detection limit = 3.3 s / S
– Where s is the standard deviation of the response
and S is slope of the calibration curve.
– The estimate of S may be carried out in a variety
of ways, for example:
• Based on the standard deviation of the blank
• Based on the calibration curve
Limit of Quantification (LOQ):
• Based on visual limitation
– Visual evaluation may be used for non-instrumental
methods but may also be used with instrumental methods.
– The quantification limit is determined by the
analysis of samples with known concentrations of analyte
and by establishing the minimum level at which the analyte
can be quantified with acceptable accuracy and precision.
• Based on signal-to-noise
– This approach can only be applied to analytical
procedures which exhibit baseline noise.
– Determination of the signal-to-noise ratios is
performed by comparing measured signals from samples with
known low concentrations of analyte with those of blank
samples an establishing the minimum concentration at which
the analyte can be reliably quantified. A typical signal-to-
noise ratios is 10:1.
• Based on the standard deviation of the response
and the slope
– The quantification limit = 10s / S
– Where s is the standard deviation of the response
and S is slope of the calibration curve.
– The estimate of S may be carried out in a variety
of ways, for example:
• Based on the standard deviation of the blank
• Based on the calibration curve
| Is This Answer Correct ? | 21 Yes | 4 No |
Answer / dr.praveen choubey
by HPLC LOD @ LOQ calculated by S/N ratio that is the best
method.LOD s/N NLT 3 and LOQ S/N NLT 10 and also calculated
by visual evalution.
| Is This Answer Correct ? | 11 Yes | 2 No |
Answer / nikhil gandhi
Based on ICH Guideline there are many method given to
identify LOD and LOQ, but most useful method is slope and
residual standard deviation is used.
LOD : 3.3 X sigma/ Slope
LOQ : 10 X sigma/ Slope
Sigma : Residual Standard Deviation
| Is This Answer Correct ? | 12 Yes | 4 No |
Answer / rakesh p. kotkar
Based on standard deviation of the response and the slope
obtained will be calculated as:
Limit of Detection (LOD):
DL= 3.3(ASD/Slope)
Limit of quantification(LOQ):
DL=10(ASD/Slope)
| Is This Answer Correct ? | 10 Yes | 3 No |
Answer / pruthvi
why we have to take 3 for LOD and 10 for LOQ? why not other numbers?
| Is This Answer Correct ? | 6 Yes | 0 No |
how do you perform reverse enginearing study for drug containing sodium chloride sodium citrate sodium bicarbonate
How to know the estimated LOQ concentration in ppm
What is differences between the Instrument and Equipment
Is sonication essential in solubility testing of a drug substance?
in which situation ion pair require to use?
Whya the need of Specificity to do in Analytical Method Validation ?
full name of OOS & OOT ? Dercribe it with details ? why this is needed in pharma industry ?
why salisylic acid not using now days for disdolution calibration?
4. Describe the operation of the Craig apparatus. Chapter.3 Equilibrium Processes in Separations
what is stability study ? why it is needed and how itis done
how to calibrate AAS ? give me the procedure?
why do u get different drug release while doing calibration of dissolution test appratus 1& 2(paddle & basket)?
1 Answers HDFC, Medrich, Twilight Litaka Pharma,