Golgappa.net | Golgappa.org | BagIndia.net | BodyIndia.Com | CabIndia.net | CarsBikes.net | CarsBikes.org | CashIndia.net | ConsumerIndia.net | CookingIndia.net | DataIndia.net | DealIndia.net | EmailIndia.net | FirstTablet.com | FirstTourist.com | ForsaleIndia.net | IndiaBody.Com | IndiaCab.net | IndiaCash.net | IndiaModel.net | KidForum.net | OfficeIndia.net | PaysIndia.com | RestaurantIndia.net | RestaurantsIndia.net | SaleForum.net | SellForum.net | SoldIndia.com | StarIndia.net | TomatoCab.com | TomatoCabs.com | TownIndia.com
Interested to Buy Any Domain ? << Click Here >> for more details...


How to check vortex flow meter is in good condition while
we do not having calibration setup?

Answers were Sorted based on User's Feedback



How to check vortex flow meter is in good condition while we do not having calibration setup?..

Answer / krishna reddy

Let me know first which make you have.

Instrument have min and max ranges on its name plate.
So wiht multimeter you can check 4-20ma of flow meter.It
must be 4 ma if flow is zero and 20ma if flow is max range.
supose your flow mete got 100L/hr.
for 4ma flow should be 0 L/hr
8ma - 25L/hr
12ma - 50L/hr
16ma - 75L/hr
20ma - 100L/hr

Like this you can check.

Is This Answer Correct ?    10 Yes 3 No

How to check vortex flow meter is in good condition while we do not having calibration setup?..

Answer / amit

U can do the cold test by checking the sensor.It has piezo
sensors in symetrical manner. U can check both piezo
sensor by checking the capacitance. If capacitance is
accordance with size means sensor is OK. same way u can
check temp and pressure sensor(by checking the resistance)
if it has.This way u can check the healthiness of the
sensor. If u want to check the response of sensor, u can
give little vibration to the sensor by hammering on it and
check the response on transmitter. U cannot directly check
the calibration for flow on field itself.

Is This Answer Correct ?    1 Yes 0 No

How to check vortex flow meter is in good condition while we do not having calibration setup?..

Answer / tuhin

@soumil: no hp and lp side exist in vortex flowmeter. Vortex
principle is different.

Is This Answer Correct ?    0 Yes 0 No

How to check vortex flow meter is in good condition while we do not having calibration setup?..

Answer / saumil

In this meter hp or lp any one side we isolate line valve and check zero if it is ok then meter is ok

Is This Answer Correct ?    1 Yes 3 No

Post New Answer

More Instrumentation Interview Questions

Is it ok, to use non-contact type Radar type LT, on a tank that is half circular from the bottom side? If yes the, isn't the waves gonna bent in other directions? 

0 Answers   Process Control Technologies,


why r using 250 resistances

6 Answers   Sabic, Samsung,


what is the input of digital RPM indicator during the calibration time?

1 Answers  


1. what is 3 point calibration & 5 point calibration? 2. If we use hart communicator to set the LRV & URV of a smart transmitter, did we calibrate the transmitter? why? 3. Make a complet list if meterials needed to properly connect a pressure gauge to a steam pipe using 1/2 inch stainless steel tubing. The gauge connections is 1/2 inch NPT-M and the steam pipe tap has a 1/2 inch female coupling, compression fittings are use. 4. What are the tools in loop checking? 5. What would the ISA symbol of a square root extractor look like? 6. What would you do to completely prepare for the configuration? 7. What safety measures must you undertake when doing configuration? 8. what are the two most commonly use programming languages used when we configure a PLC? 9. What would you do if you accidentally skipped a step in the configuration procedure? 10. How would you know if the configuration were sucessful? 11. Name at least 3 selections available when we configure the burnout protection of a controller?

3 Answers   EI Electrotechnical Institute,


What is the Differance between Graphics And SCADA Graphics? SCADA Graphics have less access?

3 Answers   Honeywell, Thermax,


HOW TO CALCULATE WAKE FREQUENCY FOR THERMOWELL???WITH OUT SOFTWARE..

1 Answers  


Why would an Engineer consider combining FEEDBACK and FEEDFORWARD control in some applications?

4 Answers   Jindal,


we are using centum vp as monitoring with Dell precision work station T3600 but some times PC(s) are restarted automatically . how we can solve this problem? thanks

0 Answers   Oil India,


Can a pneumatic control valve work without positioner?

4 Answers   SMC,


what is ILD

1 Answers  


what is Green field /brown field exposure in chemicals /oil /petrochemicals /refineries /paints /speciality chemicals /gas /FMCG /glass industry

1 Answers  


Can anybody explain about wellhead control system.

1 Answers  


Categories
  • Civil Engineering Interview Questions Civil Engineering (5086)
  • Mechanical Engineering Interview Questions Mechanical Engineering (4456)
  • Electrical Engineering Interview Questions Electrical Engineering (16639)
  • Electronics Communications Interview Questions Electronics Communications (3918)
  • Chemical Engineering Interview Questions Chemical Engineering (1095)
  • Aeronautical Engineering Interview Questions Aeronautical Engineering (239)
  • Bio Engineering Interview Questions Bio Engineering (96)
  • Metallurgy Interview Questions Metallurgy (361)
  • Industrial Engineering Interview Questions Industrial Engineering (259)
  • Instrumentation Interview Questions Instrumentation (3014)
  • Automobile Engineering Interview Questions Automobile Engineering (332)
  • Mechatronics Engineering Interview Questions Mechatronics Engineering (97)
  • Marine Engineering Interview Questions Marine Engineering (124)
  • Power Plant Engineering Interview Questions Power Plant Engineering (172)
  • Textile Engineering Interview Questions Textile Engineering (575)
  • Production Engineering Interview Questions Production Engineering (25)
  • Satellite Systems Engineering Interview Questions Satellite Systems Engineering (106)
  • Engineering AllOther Interview Questions Engineering AllOther (1379)