In my analysis by HPLC,In sample main peak and placebo
peak are eluting with shorter resolution ( almost look like
doublet). how can i get max resolution ? Is changing
gradient method is usefull?
Answers were Sorted based on User's Feedback
Answer / wasim shaikh
You can try one or more of the following:
1) Changing the column type (c8 to c18, etc.)
2) Changing the Mobile phase composition (increase buffer
etc.)
3) Decrease the particle size on column (from 5 to 4, 3
etc.)
4) Increase the column length
5) Decrease the flow rate
6) If all of the above fails, then as you suggested,
gradient method can be used to separate the two peaks.
P.S. : Gradient method can also be used as an alternative
to the above methods and not necessarily the ultimate
solution.
| Is This Answer Correct ? | 32 Yes | 1 No |
Answer / naveen
U CAN DECREASE THE FLOW RATE
AND INJECTION VOLUME
| Is This Answer Correct ? | 1 Yes | 1 No |
Answer / gaurav
You can increase the injection volume.Changing gradient
method is not required.
| Is This Answer Correct ? | 2 Yes | 14 No |
While performing TOC sst analysis Zero shift disabled & sample analysis zero shift enabled why?
why using buffers in hplc inteed of water
in each media we require to use SLS? how to proceed?
what is mean by nitrosamine impurities?
Why dissolution test is not performed in all of the products
What is cut off value of solvent used in hplc
How to choose Normal phase or Reverse Phase HPLC method Conditions for given sample?
[3/30, 13:29] Manoj P Venkatpurwar: How hplc column selection according to structure? How mobile phase buffer selection on molecule structure?
what is stability study ? why it is needed and how itis done
How do you give training to a new chemist
In the test for control of absorbance in UV calibration why do we use only potassium dichromate and what is purpose of taking a specified amount of 57.0-63.0mg?
8 Answers Aurobindo, Cipla, FDC,
From peak purity spectrum, where or at which point peak purity angle & purity threshold is checked?