what is peak purity ??
Answers were Sorted based on User's Feedback
Answer / ch.h.n.prasad reddy
Peak purity in HPLC is nothing but SPECTRUL PURITY not
Chemical purity. In HPLC Each peak is combination of many
data points. Each data point is having SPECTRUM. In peak
purity calculation each data point SPECTRUM is compared
with SPECTRUM of Peak apex Data point. In this we can
compare deferent region data points with Peak apex Data
point SPECTRUM. At that time we consider Base line noise,
Solvent noise etc.... we can select wave length range
also....
Always Purity angle is less than threshold then only Peak
Purity can pass.
Best Example: Isomers
Both isomers can show same spectrum but chemically they
are different. In this case if we get single peak also our
Peak Purity can Pass because our Peak apex Data point
SPECTRUM and deferent regions in that Peak data points
SPECTRUM is SAME.
Thanks
CH.H.N.PRASAD REDDY
RESEARCH SCIENTIST
prasadreddy19@gmail.com
Is This Answer Correct ? | 104 Yes | 10 No |
Answer / saran
In PDA Detector, Purity angle value is less than the threshold
Is This Answer Correct ? | 48 Yes | 14 No |
Answer / umesh patel
IT IS DEFINED AS PURITY OF A PERTICULAR PEAK THAT IT IS NOT
COELUTED WITH OTHER IMPURITIES
PURITY=RATIO OF ABSORBANCE B/W THE ROOT OF UP SLOPE AND
APEX/RATIO OF ABSORBANCE B/W THE ROOT OF DOWN SLOPE AND APEX
IF THE PURITY=1, PURE PEAK
Is This Answer Correct ? | 51 Yes | 20 No |
Answer / ramakrishna
to identify whether the particular peak is single or
combination of peaks.
if angle is less than threshold the peak is pure.
if angle is more than threshold the peak is impure.
Is This Answer Correct ? | 36 Yes | 14 No |
Answer / g.venkata suresh
Peak purity is nothing but purity angel should be less than purity threshould
Purity angel is a measure of the spectrul hetrogeneity of the peak based on the spectrum overall the peak using spectrum constract angel.
Spectral constract angel is defined as the shape difference between two spectra.
Purity threshould is a minimum purity of the reference standard in the formulation.
Is This Answer Correct ? | 1 Yes | 2 No |
Answer / rashid ansari
Dilute o.1 ml phosphoric acid to 100 ml with water
Is This Answer Correct ? | 0 Yes | 2 No |
Answer / kuchela rao athota
spectra are used to calculate the purity level of the peak,
or to construct the similar curves.
Similarity/Threshold
Ln(Similarity)
Similairty/Threshold ratio
Is This Answer Correct ? | 15 Yes | 19 No |
Answer / jilani
Peak purity means 100% purity of drug. SPECTRUL PURITY not
Chemical purity.
Is This Answer Correct ? | 3 Yes | 9 No |
Answer / anil kumar
Peak purity we have to check through the PDA detector. It
should be NLT99%.Peak purity we can check in 3D from uv to
visible range.
Is This Answer Correct ? | 15 Yes | 32 No |
Answer / ganesh bhavar
peak purity means 100% pure peak of drug
Is This Answer Correct ? | 10 Yes | 41 No |
if peak get problamatic then what require to do?
why we get only one peak in assay by hplc but in rs by hplc get more peaks
what is stability study ? why it is needed and how itis done
In gas chromatographic analysis sometimes we use headspace, why it is required instead of manually injection?
pda detector mechanism?
1 Answers DKR, Teva Pharmaceuticals,
In dissolution test s1 stage we are getting one bowl 40 % out of 6 bowls what is the root cause you find and how proceed next step? ( Q is 80%)
what is the procedure for cleaning of lenses of hatr accesory of ftir instrument?
what is mean by assay?
Why Deuterium lamp is used as UV light Source for UV.
In case of HPLC coloums what is the difference Between C8 & C18 coloums & how can we select difference type of coloums like Inertial ODS , Hypersil BDS , ODS etc.?
in what basis wavelength will select for a compound at initial method development trials
If we have 120 minutes run time method for HPLC and you wan to decrease the Run time so what should you do??