Answer / falguni patel
Dead volume: It comprises the volume of HPLC system between
the point of injection to the point of detection, excluding
the part of column that contains packing.
Dwell volume: It includes the volume between pump and
column. in low pressure gradient it may be 1 mL & in high
pressure gradient it is to less.
| Is This Answer Correct ? | 14 Yes | 2 No |
Answer / aks
Dead Volume:Volume between injector to detector excluding
column volume (and also called as extra column volume)
Dwell Volume:Volume form the point of mixture to the column
entry (Also called as system delay volume)
| Is This Answer Correct ? | 10 Yes | 0 No |
what is the diffrence between absorbance and absorption in the beer,s lamber formula ?
Why only hydroxy naphthol blue indicator is used for standardisation of 0.05M EDTA solution instead of solochrome Black T or Euriochrome Black T indicator which is used for all sample analysis with 0.05M EDTA solution?
In GC and HPLC ,how we can decide which of following method will be suitable or use for quantitation of chromatogram at the time of method development. 1.Area Normalisation 2.External standard method 3.Internal standard method
What is Signal to noise ratio?
Why noise test is performed at 254 nm in calibration of HPLC system?
What is the reason of RT change in new column v/s old column in HPLC
what is the differance between Analytical and semi micro balance?
What is the importance of rrf in rs by HPLC?
If we have 120 minutes run time method for HPLC and you wan to decrease the Run time so what should you do??
What are usp limits for total impurities,unknown impurity
how to set sample and standard concentration in RS method?
Difference between uv detector nd RI DETECTOR? wt s the difference between OVI Solvents nd residual solvents incase of GAS CHROMATOGRAPHY? nd wt s its detection limit?