what are the different methods of analysing related
substances test,what are LOD AND LOQ .HOW TO REPORT
IMPURITIES WHEN LOD AND LOQ LIMITS R SPECIFIED
Answers were Sorted based on User's Feedback
Answer / ananthan chandrasekar
LOD-Limit of detection.It is the lowest amount of the
analyte that can be detected and not necessary to be
quatitated.
LOQ-Limit of Quantitation. It is the lowest amount of the
analyte that can be detected and quantified.
If u get the impurity value below LOD, impurity result can
be declared as BDL(Below detection limit)
If u get the impurity value morethan LOD value but lessthan
LOQ value,impurity result can be declared as BQL(Below
Quantitation limit)
Only those impurities which are exceeding LOQ value to be
declared as per calculated value and this value canbe
included for Total impurity calculation.
For reference, See the example
Impurity LOD LOQ Calculated value Result
A 0.002% 0.006 % 0.001% BDL
B 0.002% 0.006 % 0.003% BQL
C 0.002% 0.006 % 0.008% 0.008%
D 0.002% 0.006 % 0.009% 0.009%
Total impurity = 0.008% +0.009 % =0.017%
| Is This Answer Correct ? | 58 Yes | 3 No |
Answer / hari prasad
LOD means Sample concentration whereever gets the lowest
detection and there sample has not quantified.
LOQ means Sample concentration whereever gives lowest
quantified detection and there sample has quantified.
The LOD&LOQ concentration where the precission %RSD would
have 'not more than 30% and 10%' respectively, those are
the finalised LOD and LOQ limits to report.
Now impurities can report based on established LOD&LOQ
limits i.e. Not detected or below LOQ.
| Is This Answer Correct ? | 15 Yes | 3 No |
In Karl fischer titrator what medium we can choose to analysing aldehyde and ketone samples by using KF reagent
What is the defination of Bulk density?
How to establish relative response factor for hplc. Why it is required?
How decided 10 ppm standard solution in Swab, rinse method.?
what does pH stand for?
i want to know about headspace and interview question on headspace
1 Answers Srikem Laboratories,
why urasil use in hplc column perfomance?
why potassium dicromate use in calibration of UV VIIBLE SPECTROCOPY ?
Why Holium cholorate used in UV Spectro photometer?
How will you separate Colorless compound Using Column Chromatography?
4 Answers Delhi University, VPHS,
why we cannot analysis below 200nm in u.v?
does water content by KF will calculate only external moisture absorbed or the ligand ? (for ex copper sulphate pentahydrate)