How to calculate LOD & LOQ in HPLC Validation Method
Answer Posted / m
There are three different method indiated in ICH guidelines
for determination of LOD and LOQ.
1 Visual evaution method - For non instrument and sometimes
for instrumental also.
2 Based on residual standard deviation of the response and
the slope
obtained will be calculated as:
Limit of Detection (LOD):
DL= 3(SD/Slope)
Limit of quantification(LOQ):
DL=10(SD/Slope)
3 Signal to noise ratio method
But residual standard deviation method is more reliable
method for determination of LOD and LOQ value
| Is This Answer Correct ? | 138 Yes | 17 No |
Post New Answer View All Answers
in stress study if your api not soluble in hcl naoh h2o2 then what require to do?
how can give the expiry period and restadardisation of volumetric solution
how you start the solubility study?
4. Describe the operation of the Craig apparatus. Chapter.3 Equilibrium Processes in Separations
what is mean by ambient temperature?
why we are using hexane in calibration of number of drop per mL
how you establish the LOQ?
why glutent are detected in the rice cereal baby food product even manufacturer claimed that they are using rice and milk only?we have using ELISA to do the test,and rice supposed not containing any glutent,rite?We already repeat the test so many times and it still detected.just wondering where the glutent came from?
CAN WE DETERMINE THE PEAK PURITY IN GAS CHROMATOGRAPHY ANALYSIS?
In IR spectra, generally a molecule vibrates without apply emr or ir light?
If suppose 10 methods of dissolution given in pharmacopoeial for single content product then which method out of 10? or all 10 require to follow?
cefoperozone and sulbactam inj. hplc test method
in dissolution why pool sample needed? in which type of drug pool sample need?
how to get accurate result for Residue on evaporation in purified water
Why we check moisture,ash & AIA in product