Why noise test is performed at 254 nm in calibration of HPLC
system?
Answers were Sorted based on User's Feedback
Answer / paresh jobanputtra, piramal
Earlier fixed wavelength detector is mostly used i.e 254 nm
because almost all Benzene ring containing compounds absorbs uv light
at 254 nm. i think due to this reasons only.
Regards,
Paresh Jobanputtra
| Is This Answer Correct ? | 18 Yes | 1 No |
Answer / tejas shah
At 254nm none of the solvent or its impurities show significant interference so the noise will be recorded of pure UV light. i.e. detector noise
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / jayesh patel
Generally all organic compounds have absorption at this
wavwlength.
| Is This Answer Correct ? | 3 Yes | 6 No |
What is the principle of hplc detectors like.. UV detector,flurosence detector, RID?
why using factor 1.00639 in burette calibration of kf
In acid base titration, after complet consumption of HCl by NaOH, addition of next drop of NaOH will give pink colour with phenolpthalein indicator. Why this pink colour appear? Why not blue, black, green or anything elese?
Why are use silicon oil mr?
I have compare C2H2-air and C2H2-N2O flame AAS on determination calcium. Both use same range of std to plot calibration curve. (2-6ppm) When i measure the sample with phosphate, KCl and LaCl, C2H2-N2O flame give false positive result, around 0.5ppm. When i measure the sample with phosphste, KCl and EDTA. C2H2-N2O flame also give 0.5ppm false positive. But both above mentioned sample would not give false positive when measured by C2H2-air flame. What is the reason?
In C8,C18....etc have 4&5 etc... microns Columns is there like that how to separate Chiral columns as per Micron size?
In partical size by malven,which parameter is generlly change for development of dry and wet method.
what is classification of elemental impurities? what is risk assement in elemental impurities?
who is the father modern liqiud chromatogrphy?
3 Answers Amneal Pharmaceuticals, Car Chem, Cipla,
what are the sources in ELSD,PDA, RI and UV Detectors?
for which product require to do content uniformity? what is limit of cu?