If we have 120 minutes run time method for HPLC and you wan
to decrease the Run time so what should you do??
Answers were Sorted based on User's Feedback
Answer / pankaj shrivastava
There are some factors and parameters by which we can
decrease the run time as....
1. by increasing flow Rate
2. by increasing lightly Column oven temperature.
3. by increasing mobile phase strength.
3. by using short length column.
| Is This Answer Correct ? | 28 Yes | 3 No |
Answer / priyanka.r
1)By changing the mobile phase composition of organic solvent(for ex:acetonitrile),then RT will shift.
2)by changing the flow rate, length of the column, temperature.
3)the next option is better choose some advanced Technic like UPLC.
| Is This Answer Correct ? | 13 Yes | 1 No |
1) By increase the mobile phase ratio of organic solvent,
2)By increasing Flow rate, 3)Go for Gradient Progrom 4)
Reduce the column Length And Particle size 5)By Changing the
pH of the mobile phase RT will shift
| Is This Answer Correct ? | 5 Yes | 0 No |
Answer / nirav
By addition of organic solvent in mobile phase, By changing
the flow rate, By changing length of the column.
| Is This Answer Correct ? | 3 Yes | 3 No |
Answer / ra
You must try for UPLC.Or otherwise you have to do hard work on the method.
| Is This Answer Correct ? | 3 Yes | 8 No |
Answer / santhu
BY ADJUSTING THE PH OF THE MOBILE PHASE WITH BUFFERS
| Is This Answer Correct ? | 1 Yes | 7 No |
how to calculate corelation coefficient in G.C Calibration? what is the calculation part?
wat is the role of michelson interferometer in FTIR and He-Ne lase.detailed answer required in simple language
Why do we use 500ppb sucrose and 500ppb 1,4 benzoquinone in TOC calibration?
When ever we perform system suitability,inject the corresponding standard 6 times why? is thair any reson
In UV-VIS spectrophotometer:My diluent is methanol even though i corrected base line for methanol,In my sample scan i found methanol peak.what is the reason?
Tell me about analytical method validation in QC
AT WHAT CONDITIONS WILL YOU ANALYSE A SAMPLE,WHICH ARE PRESERVED AT OTHER THAN THE AMBIENT CONDITIONS?(ie COLD STORAGE SAMPLES BELOW 20* CENTIGRADE?
which functional group we can detect in IR?
Can anybody tell me how to measure Redox potential in an aerobic matrix, for e.g., soil etc.?
Why we don't add Hcl, HNO3, H3PO4 in residue on ignition, why we add H2SO4
why we use prednisone tablet in chemical callibration of dissolution apparatus?
6 Answers Cipla, Mylan, Umedica Laboratories,
Why Holium cholorate used in UV Spectro photometer?