how to hplc calibrated
Answers were Sorted based on User's Feedback
HPLC is calibrated by the following parameters:
1) Pump calibration
2) Detector Linearity
3)Injector Precision and carryover test
4) Wavelength accuracy by using caffeine and pyrene solution
5) Determine the column oven temp. and sample compartment temp
6) Gradient test
7) Detector noise
Is This Answer Correct ? | 128 Yes | 13 No |
Answer / ajinkya p.
In HPLC there must calibrations for Pump, Detector,injector
Pump :- Should calibrated by checking flow accuracy test, Pressure holding test.
Detector :- Should calibrated by checking Linearity test, Detector wavelength accuracy test
Auto sampler :- Should calibrated by Injector Accuracy test,
Carryover test, Injector precision test, carry over test
Is This Answer Correct ? | 48 Yes | 2 No |
Answer / kanakadurga
calibration parameters are
1.Flow accuracy
2Temperature accuracy
3Wave length accuracy by using caffeine and erbium perchlorate
4Linearity of detector and injector
5.Carry over test
6.Drift and noice
Is This Answer Correct ? | 37 Yes | 6 No |
Answer / b.ganesh
calibration parameters are
1.Flow accuracy
2Temperature accuracy
3Wave length accuracy by using caffeine and erbium perchlorate
4Linearity of detector and injector
5.Carry over test
6.Drift and noice
7.Gradient progromme accuracy
8.injector precision
Is This Answer Correct ? | 3 Yes | 1 No |
Why dissolution test apparatus calibration with salycilic acid tablets was stopped by FDA&USP from the year of 2009?
In which situation we require to analytical method validation of excipient?
in which situation require to take incident in validation?
why using the ph 4,7,9.2?
What is the accrptance criteria in RSD for RS method precision on basis of impurity percentages?
What is the difference between Primay and Secondry Standard ?
6 Answers College School Exams Tests, Neuland,
how can standardized of HCl of 0.1 N
what type of compounds do we see lod @ wat type compounds do we see m.c
Is sonication essential in solubility testing of a drug substance?
if your impurity coeluting with each other in that situation how require to set specification? is it acceptable?
what is the general mechanism of high performance liqiud chromatography?
7 Answers Ajanta Pharma, Amneal Pharmaceuticals,
If i performing one analysis on C8column then the same analysis done on C18 column then what will happen about Rt;Resulution;Peak shape Explain ?