Answer / rakesh
1.0 CONTROL OF ABSORBANCES
PROCEDURE: REPARATION OF POTASSIUM DICROMATE SOLUTION: Dry
K2Cr2O7 at 130oC for 2 hours. Weigh accurately about the
quantity of sample equal to 60mg into 1000ml volumetric
flask. Add 0.005m Sulphuric acid to dissolve and dilute
with the same solvent.
Find out the absorbances of (A 1% 1cm) above said solution
at wavelengths of a 235nm, 257nm, 313nm and 350nm.
Calculate the (1% 1cm) value using the following expression.
Absorbance X 1 X 1000 / Weight of K2Cr2O7 in g X 100
Check the value for each wavelength with the tolerances
given four wavelengths.
PREPARATION OF REAGENT:
(i) Dilute 3ml of concentrated sulphuric acid and make up
100ml with distilled water.
(ii) Dilute 20ml to 2000ml with distilled water.
2. LIMIT OF STRAYLIGHT:
The absorbance of a 1.2-% w/v solution of potassium
chloride at a path-length of 1cm should be greater than 2.0
at about 200nm when compared with water as reference liquid.
1.2% POTASSIUM CHLORIDE SOLUTION:
Take 1.2 g of Potassium chloride previously dried at 105oC
for two hours into the 100ml volumetric flask. Add water to
dissolve and make up to volume with water.
3. RESOLUTION POWER:
Record the spectrum of a 0.02% v/v solution of Toluene in
Hexane UV. The ratio of the absorbance at the maximum at
about 269nm to that at the minimum at about 266nm is not
less than 1.5 unless other wise specified in the monograph.
0.02% v/v TOLUENE:
Take 0.02ml of Toluene into the 100ml volumetric flask. Add
Hexane to volume.
| Is This Answer Correct ? | 97 Yes | 8 No |
Answer / upk
UV-VIS spectrophotometer calibration is many parameters
which is
1)control of wavelength
2)control of absorbance
3)wavelength accuracy
4)straylight accuracy
5)resolution
6)resolution power &
7) Io Line Flatnes test
| Is This Answer Correct ? | 55 Yes | 22 No |
Answer / nitesh pal
in above answer also added 430 nm in control of wavelength.(acc to ipm dilution is about 60 mg to 1ooml (same solvent)
| Is This Answer Correct ? | 21 Yes | 8 No |
Answer / devender justa
Internal Calibration ,external Calibration, Baseline
Flatness
| Is This Answer Correct ? | 8 Yes | 4 No |
How to define samples weight in moisture content
full name of OOS & OOT ? Dercribe it with details ? why this is needed in pharma industry ?
can i use hplc detector to uplc and why?
In acid base titration, after complet consumption of HCl by NaOH, addition of next drop of NaOH will give pink colour with phenolpthalein indicator. Why this pink colour appear? Why not blue, black, green or anything elese?
In Force degradation Study Compound is not degrade in any condition , also in higher stress what can i do ?
Defination of ZEEMAN EFFECT in NMR ?
which solvent is highest poler acetonitrile dimethyl formamide methanole tetrahydrofurane buffer solution
we are performed SOR for a particular product the limit is -6 to -10.We face particular bathes it is not meeting the specification.we performed diffrent instrument diffrent analyst we are getting diffrent diffrent results.solvent is dimethyl sulfoxide what could be the reasion
what is dwell volume in HPLC?
2 Answers Aurobindo, Cadila, Strides, SUN, Sun Pharma,
What is dose dumping? why require to do?
For LOD and LOQ determination which method is accurate? (slopemethod or s/n ratio method)
What is shaking level in GC?