same sample on two preapration gives differnt result during
hplc analysis
Answers were Sorted based on User's Feedback
Answer / hanimi
Before conclusion we have to review the data. investigation
has to done on test method and on experiment.
if there is no problem in experiment. It may have two
reasons
1. Analytical method is not suitable to get the the
consistant results
2. Sample is not uniform
| Is This Answer Correct ? | 10 Yes | 0 No |
Answer / reet
there may be any one of the reasons:
1. Air bubble in the injector needle (needle wash 2-3 times0
2. carry over might be there.
3. random variation in the flow rate
4. Integration problem (chk the processing method)
5. Sample itself may a problem !
i. solubility issues
ii.sample(generally proteins may be sticking to the
walls of the sample vial and also may be getting
irreversibly stuck on the surface of the column)
iii. temperature change.
| Is This Answer Correct ? | 5 Yes | 0 No |
What is difference between NDA & ANDA ?
what are risk assement in the analytical qbd?
what is peak purity? how may types of methods to determine peak purity?
what is use in calibration of conductivitimeter?
1 Answers Claris Lifesciences, SRF,
principle of UV , PDA , RI detector in HPLC
HOw do i calculate the number of theoretical plate on a HPLC chromatography
how you determine moist.of NaOH sol or NaOH flacks
Why do we use sodium salicylate for Calibration TLC Cabinet U.V Lamps
Why chromic acid wash is not recommended by USP for glassware cleaning?
reasons for negative peaks in chromatography
What is the difference of time between two consecutive injections in HPLC?
2 Answers Nectar Lifesciences,
Why do we use 500ppb sucrose and 500ppb 1,4 benzoquinone in TOC calibration?