Why DP transmitters are subjected to high static pressure
test of 100 bar. Though they will be used for very small
pressure range
Answers were Sorted based on User's Feedback
Answer / hiren
dp transmitter is used to measure DP but the pressure
applied across diaphargm is actual process pressure. so if
any one get fail then transmitter should not fail.
| Is This Answer Correct ? | 10 Yes | 1 No |
Answer / komal
Static Pressure is done on DP transmitter because of check
if the transmitter can withstand the static pressure acting
due to Line Pressure when in process. The Line pressure can
be 100 bar but the DP that is measured on transmiter can be
as small as 1000 MM H2O.
| Is This Answer Correct ? | 6 Yes | 1 No |
Answer / mhakave rajesh
it's span of measurement may be less eg 0- 5 bar,but
pressure acting on single side either of hp or lp (i.e.
process side pressure) may be high, so to check zero of tx
at such operating pressure it is........ so and so.
| Is This Answer Correct ? | 2 Yes | 1 No |
Answer / b.s.m.raju
This is DP transmeter high static preesure of 100 bar.
Small range will be used 10 mbar. but span range very low
Accuracy will be poor.
| Is This Answer Correct ? | 1 Yes | 2 No |
what is effect of specific gravity in differential level transmitter , explain formula related with this .
0 Answers Flour Tech Engineers (P) Ltd,
What is the difference between Error and Offset.
Is there’s power on the bus (if so, state whether on the signal bus, or on a separate in the following: a) HART b) Profibus DP c) Profibus PA d) DeviceNet e) Foundation Fieldbus
what is the dp across a orifice plate & how much is permanent pressure across an orifice plate.
IF PLC POWER FAILURE THEN WHAT HAPPEN?
Why is the orifice tab provided?
I am having a Phoenix digital to modbus converter. The o/p of the same is connected to flow computer FB407. But the digital signals are updated only after one minute in the flow computer. what may be the cause of delay in communication? Flow computer is having a scanning period of less than 1 sec.
Area classification, Equipments, Gas groups, temperature classification, Ingress protection, Fire Extinguishers.
Why do we have earthing cable in branch cable and shield in the same cable?
0 Answers Process Control Technologies,
What is meant by cold float of compressor shaft.
i have one absolute pressure transmitter for the range is -400mmhg to 400mmhg. If i do the zero trim at atmosphere the transmitter should be give 4ma or 12ma? which one is correct?
Tyes of interface level measurement and how to use that technique for interface level measurement,
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)