why used three wire in RTD.
Answers were Sorted based on User's Feedback
Answer / preeti
three wire connections can eliminate connection lead resistance effects from measurements
Is This Answer Correct ? | 15 Yes | 4 No |
Answer / gomsy
third wire is used as lead wire compensation i.e it is
connected in parallel,hence effective resistence decreases
and will be helpful to calculate higher range of tempreture
Is This Answer Correct ? | 13 Yes | 4 No |
Answer / santosh r apre
three wire connections can eliminate lead resistance from
mesurements.
Is This Answer Correct ? | 8 Yes | 2 No |
Answer / sam.r
Most RTD’s incorporate a third wire with resistance R3.
This configuration provides one connection to one end and
two to the other end of the sensor. Connected to an
instrument designed to accept 3-wire input, compensation is
achieved for lead resistance and temperature change in lead
resistance. This is the most commonly used configuration.
If three identical type wires are used and their lengths are
equal, then R1 = R2 = R3. By measuring the resistance
through leads 1, 2 and the resistance element, a total
system resistance is measured (R1 + R2 + RE ). If the
resistance is also measured through leads 2 and 3 (R2 + R3),
we obtain the resistance of just the lead wires, and since
all lead wire resistances are equal, subtracting this value
(R2 + R3) from the total system resistance (R1 + R2 + RE)
leaves us with just RE, and an accurate temperature
measurement has been made.
Is This Answer Correct ? | 5 Yes | 1 No |
Answer / seshu
The connecting wires long,resistance of wire changes as ambient temp fluccuates.to eleminate this problem we use 3 wire rtd
Is This Answer Correct ? | 9 Yes | 10 No |
the RTD is
installed some distance away from the transmitter. Since
the connecting
wires are long, resistance of the wires changes as ambient
temperature
fluctuates. The variations in wire resistance would
introduce an error in the
transmitter. To eliminate this problem, a three-wire RTD is
used
Is This Answer Correct ? | 2 Yes | 3 No |
why 8bit microprocessor referred as 2^16?
List three types Level Gauges with their working principles
In the field and in the control room different power supplies if used How to handle the ground
What is difference between profibus and modbus communication
the square root type transmitter r used to which applications
How to Calibrate Interface Level Troll in field
What is the meaning of wet leg
In RS 500 SLC 5/03 input change one module card to another card, plz. give steps clearly.
what is cems analyzer explain with principle and drawing ?
what is Least Value of MICROMETER?
Sitrans LR460 Radar level transmitter getting level out or disturbed level during moisture present in the material ie coal in the pccb bunker in CHP.Any suggestions for this problem...????
How DP Transmitter can be calibrated for crud application in floating roof tank?