IF HPLC CALIBRATION FAILED THEN WHAT YOU DO
Answers were Sorted based on User's Feedback
Answer / natraj kumar
Example HPLC calibration : - Suppose one parameter is
failed .keep the status board for Instrument breakdown.And
raise the OOC (Out of calibration )document take the
approval from QC and QA Head. After rectification or change
the part of the Instrument, check the failed parameter of
calibration.
If it is Passed, then check randomly impact assesment of
the previous analysis during this calibration period.If
there is no impact then u can proceed for regular
usage.Total process to be complete in 30 days only.
OOC document keep the calibration file. Service report is
also keep the filing internal or External service person.
Dr.Reddys Lab
| Is This Answer Correct ? | 72 Yes | 3 No |
Answer / k.g.gupta dr.reddys
Give the Instrument for breakdown. After rectification of
the Instrument, check the whole parameters of calibrations.
If it is Passed, then check randomly impact assesment of
the previous analysis during this calibration period.If
there is no impact then u can proceed for regular usage.
| Is This Answer Correct ? | 22 Yes | 5 No |
Answer / ajinkya p.
If hplc calibration fails, first we can cross check that all parameters are correct, sample preparation correct or not, system washed proper or not, condition of column is ok or not. first we will try to solve issue ourself. but if still calibration fails then discuss issue with senior person so that if there minor problem so that we can solve it ourself by their experience. but if any major problem we can take help by customer online support calling of instrument company person. bt still calibration fails we need to contact instrument engineer for visit.
| Is This Answer Correct ? | 11 Yes | 0 No |
Answer / kuchelarao
ONCE U CAN CHECK THE PREVIOUS ANALYSIS AND IT IS IMPURITY
PROFILE AND THAT BATCH MUST BE COMMERCIAL AND ANALYSED TAHT
SAMPLE U CAN CHECK IT OUT PREVIOUS IMPURITY PROFILE BOTH
PROFILES ARE SAME THAT IS NO PROBLEM NOT MATCHED CAL THE
SERVICE ENGENEER.INFORM THE CUSTOMER......
| Is This Answer Correct ? | 10 Yes | 12 No |
[3/28, 20:52] Manoj P Venkatpurwar: how many impurities require to inject in assay specificity that how we can find out? and in Rs also how?
As per ICH related substances stability trend limit from initial to shelf life
0 Answers Medley Pharmaceuticals, Micro Labs,
How to calculate LOD & LOQ in HPLC Validation Method
14 Answers Dr Reddys, GSK GlaxoSmithKline, Reddy Labs, Reliance, UIS,
define beer-lambertslaw.
what is primary solution and secondary solution
I got 88.0 % assay for a drug in HPLC analysis for same injection i got 40.0% related substances is it possible? what is the relation between results of assay and purity?
What is similarity factor and its use in api strength?
cefoperozone and sulbactam inj. hplc test method
0 Answers Graviti Pharmaceuticals,
Why are using in Hi,zero air, Ni gass are used GC
5 Answers Aurobindo, ELBS, Vimta Labs,
why holium oxide prepared in 10% HCLO4 SOLUTION
why required specificity in valdation
USP methodology, EP methodology, IP methodology, among three if possible to use one methodology for qualify working standard to use USP, EP, IP ? Please explain...
0 Answers Srini Pharmaceuticals,