IF HPLC CALIBRATION FAILED THEN WHAT YOU DO
Answers were Sorted based on User's Feedback
Answer / natraj kumar
Example HPLC calibration : - Suppose one parameter is
failed .keep the status board for Instrument breakdown.And
raise the OOC (Out of calibration )document take the
approval from QC and QA Head. After rectification or change
the part of the Instrument, check the failed parameter of
calibration.
If it is Passed, then check randomly impact assesment of
the previous analysis during this calibration period.If
there is no impact then u can proceed for regular
usage.Total process to be complete in 30 days only.
OOC document keep the calibration file. Service report is
also keep the filing internal or External service person.
Dr.Reddys Lab
| Is This Answer Correct ? | 72 Yes | 3 No |
Answer / k.g.gupta dr.reddys
Give the Instrument for breakdown. After rectification of
the Instrument, check the whole parameters of calibrations.
If it is Passed, then check randomly impact assesment of
the previous analysis during this calibration period.If
there is no impact then u can proceed for regular usage.
| Is This Answer Correct ? | 22 Yes | 5 No |
Answer / ajinkya p.
If hplc calibration fails, first we can cross check that all parameters are correct, sample preparation correct or not, system washed proper or not, condition of column is ok or not. first we will try to solve issue ourself. but if still calibration fails then discuss issue with senior person so that if there minor problem so that we can solve it ourself by their experience. but if any major problem we can take help by customer online support calling of instrument company person. bt still calibration fails we need to contact instrument engineer for visit.
| Is This Answer Correct ? | 11 Yes | 0 No |
Answer / kuchelarao
ONCE U CAN CHECK THE PREVIOUS ANALYSIS AND IT IS IMPURITY
PROFILE AND THAT BATCH MUST BE COMMERCIAL AND ANALYSED TAHT
SAMPLE U CAN CHECK IT OUT PREVIOUS IMPURITY PROFILE BOTH
PROFILES ARE SAME THAT IS NO PROBLEM NOT MATCHED CAL THE
SERVICE ENGENEER.INFORM THE CUSTOMER......
| Is This Answer Correct ? | 10 Yes | 12 No |
why phepolphthalene giving pink color when we titrate against to NaOH Solution ,giving colrless titrating aginsiting with acids?
2 Answers Graviti Pharmaceuticals, Vimta Labs,
What is the difference between chromatographic purity and related substances?
why 6 tablets use for dissolution
1 Answers Cadila, IPCA, SUN, Teva Pharmaceuticals,
Why KBR is used in IR for preparation of pellete, why it shows zero dipole moment
use of stry light
why we use sodium salicylate in uv light calibration at TLC chamber
1 Answers Aurobindo, Lee Pharma,
Why red HgI2 added during starch preparation?
What is viod volume and peak purity in HPLC?
on saturation solubility study data how we can find out the bcs class of drug?
what is the difference between c-8 and c-18 column
wat is the role of michelson interferometer in FTIR and He-Ne lase.detailed answer required in simple language
4. Describe the operation of the Craig apparatus. Chapter.3 Equilibrium Processes in Separations