IF HPLC CALIBRATION FAILED THEN WHAT YOU DO
Answers were Sorted based on User's Feedback
Answer / natraj kumar
Example HPLC calibration : - Suppose one parameter is
failed .keep the status board for Instrument breakdown.And
raise the OOC (Out of calibration )document take the
approval from QC and QA Head. After rectification or change
the part of the Instrument, check the failed parameter of
calibration.
If it is Passed, then check randomly impact assesment of
the previous analysis during this calibration period.If
there is no impact then u can proceed for regular
usage.Total process to be complete in 30 days only.
OOC document keep the calibration file. Service report is
also keep the filing internal or External service person.
Dr.Reddys Lab
| Is This Answer Correct ? | 72 Yes | 3 No |
Answer / k.g.gupta dr.reddys
Give the Instrument for breakdown. After rectification of
the Instrument, check the whole parameters of calibrations.
If it is Passed, then check randomly impact assesment of
the previous analysis during this calibration period.If
there is no impact then u can proceed for regular usage.
| Is This Answer Correct ? | 22 Yes | 5 No |
Answer / ajinkya p.
If hplc calibration fails, first we can cross check that all parameters are correct, sample preparation correct or not, system washed proper or not, condition of column is ok or not. first we will try to solve issue ourself. but if still calibration fails then discuss issue with senior person so that if there minor problem so that we can solve it ourself by their experience. but if any major problem we can take help by customer online support calling of instrument company person. bt still calibration fails we need to contact instrument engineer for visit.
| Is This Answer Correct ? | 11 Yes | 0 No |
Answer / kuchelarao
ONCE U CAN CHECK THE PREVIOUS ANALYSIS AND IT IS IMPURITY
PROFILE AND THAT BATCH MUST BE COMMERCIAL AND ANALYSED TAHT
SAMPLE U CAN CHECK IT OUT PREVIOUS IMPURITY PROFILE BOTH
PROFILES ARE SAME THAT IS NO PROBLEM NOT MATCHED CAL THE
SERVICE ENGENEER.INFORM THE CUSTOMER......
| Is This Answer Correct ? | 10 Yes | 12 No |
In the gas chromatography what is used carrier gas and make up gas?
what is role of pH in mobile phase
3 Answers Cipla, Dhanuka Laboratories,
Why dissolution test performed on six jar?
What is the difference between Discriminating media and DPDM(Dissolution Profile with Different Media)
how do start hplc method development? how much sample required for developing the method development?
3 Answers BMS, Cipla, Indoco Remedies, Sun Pharma,
WHAT IS DIFFERENCE B/W MALTINGRANG/DSC
why you are using benzoic acid for water content determination in TEA
what is the defination of calibration & verification?
For limit test of heavy metals in BP, Method C require that the substance is ignited at a temperature not exceeding 800 °C. Why confines such the temperature?
IS THERE ANY EQUIPMENT TO CHECK AND CALCULATE THE POLARITY OF A LIQUID?
why phenolpthalin gives pink colour when we titrate aganist sodium hydroxcide & phenolpthalin gives colourless titrate aganist hydrochlric acid which is un burate
What is the Deference between Bulk density & Tap density ?
2 Answers Hikma Pharmaceuticals, Neuland, Reddy Labs,