why linerity not use in GC but use in HPLC ?
Answers were Sorted based on User's Feedback
Answer / mr.bujji reddy kanchi
Linearity can be used in GC also.
Lineariy must be evaluated in case of analytical method
validation and calibration.
| Is This Answer Correct ? | 18 Yes | 3 No |
Answer / d.sambireddy
Linearity also using in gc
In Any Analytical method validation and calibration of
GC. Linearity is one of the Importent parameter
| Is This Answer Correct ? | 4 Yes | 0 No |
Answer / jangam shyam
linearity test used in hplc and gc.this test is performed to
detect the performance of detector.if we increase the conc
of the sample vise versa the response is increased.the graph
shold be linear.
| Is This Answer Correct ? | 3 Yes | 0 No |
Answer / sandip bhasme
If you are taking % area of a componant; there is no need to
take linearity in GC, But if you have to compare standard
against sample, you need to check linearity for GC also. eg.
if you are analyzing a gas sample on FID OR TCD detector,
your detector response must be linear.
| Is This Answer Correct ? | 2 Yes | 0 No |
Answer / madhu
linearity is one of the main importent calibration parameter in GC& hplc
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / nani
Linearity should be evaluated GC.
1.Detector responce linearity
it shows your dector linea or not
| Is This Answer Correct ? | 0 Yes | 1 No |
What is difference between NDA & ANDA ?
What is the difference between chiral cel OD-H & OD-RH ?
Why pH range is 1-14 nt more than 14???
3 Answers IPCA, Lincoln Pharmaceuticals,
why peak purity not considered in GC?
4 Answers DRL, Reddy Labs, Vimta Labs,
What is the optimum period of storage of indicators?
1 Answers Orchid Chemicals and Pharmaceuticals,
How to calibrate glassware against std calibrated glassware?
iam using ph instrument lab india make.3ponit and 5point.in calibration time 3point slope will come 98.89and 96.87.that limit is 85 to 105.what basis slope will come.any calucalation between them.5point wiicome slope 99.89?why 3p0int came 2 values
In GC, What does 'Column condensation test' mean by?
why require to do water content for drug product?
what is the moisture content of Disodium hydrogen tartrate. how it can be calculted. as 16.5%
I have compare C2H2-air and C2H2-N2O flame AAS on determination calcium. Both use same range of std to plot calibration curve. (2-6ppm) When i measure the sample with phosphate, KCl and LaCl, C2H2-N2O flame give false positive result, around 0.5ppm. When i measure the sample with phosphste, KCl and EDTA. C2H2-N2O flame also give 0.5ppm false positive. But both above mentioned sample would not give false positive when measured by C2H2-air flame. What is the reason?
why we are using only lead standard solution for Heavy metals test? what is the purpose of this?