Answer / amith.patil17@gmail.com
Calibration Can be Carried out By Calibration Instruments
such As For Pressure Transmitter,level transmitter the
pressure pumps with mA in Series or Can be used simply
HEART with a Pressure pump,By increasing the pump pressure
the master guage shows the indication of the Standared
pressure Corresponing to which mA reading canbe Obtained
check for incrasing and decreasing Valus for 4,8,12,16mA If
Error Occurs then Set a Zero or Span Adjustment,Simillarly
For RTD,T/C in Furnace is Carried out.
| Is This Answer Correct ? | 16 Yes | 1 No |
Answer / ram
Calibrate to any instruments like pressure transmitter by using the druk calibrator to build the pressure depends on transmitter range and measure the 4-20 ma it's increase or decrease have a span and zero settings .0_30bar like initially 4ma,8ma,12ma,16,20ma.and using hart communicator also
| Is This Answer Correct ? | 3 Yes | 0 No |
Calibration of instruments are carried out under master equipment,they are calibrated under the "National Instrumentation Laboratory".
I can not tell u about the detail because it is under rule But i will explain it with example:
To calibrate a temperature Gauge-
The master equipment use is "Fluidless temperature Calibrate".
it work With out fuel so it is called Fluidless Temperature Calibrator the only thing u have to do is Put the End nod into the master equipment and initially set the temp. to 0degree cent. and step up to 30degree cent. with this see the reading proportionally in the Gauge no need to calibrate if it give +-0.2 reading. if it give deviation of +-5 then adjust the span and zero to set point result we been seen these is the method of calibration of temperature gauge.
Other example of master equipment are microprocessor based pressure calibrator,Macloid gauge etc.
| Is This Answer Correct ? | 0 Yes | 3 No |
hi this is noor in my work spot, i arranged THERMOCOUPLE in the place of RTD. (in FEED WATER tank). can i use T/C or not?
what is differential pressure transmitter and transducer? how to work?
which type of filter gives the constant phase characteristics?
what is the working principle of hydra step(electronic level indicator)?
2 Answers Bengal Energy, Lanco,
What are the cable come out from the paddle type flow sensor.and how is further connections
hello can yoy pls send me previous papers of jindalpower and steel ltd.....'m from instrumentation..... mani1048@gmail.com
i have calibrate rosemount make level transmitter , wht can i do ? my tank capacity is 30,000 liter
Please tell me details about Sox,Nox,SWAS analyser????
Why trim span ? How?
give eg of how I/P is used in control loops
Make a ladder diagram on LSL valve will close, on PSL valve will open and start the motor
State four different types of flow measurement, for gas flow measurement MMSCFD is the common unit expand the unit and specify significance of word S in the unit.
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)