Describe how to calibrate various types of instruments and
transmitters using ISA industrial consensus standard
techniques.

Answers were Sorted based on User's Feedback



Describe how to calibrate various types of instruments and transmitters using ISA industrial conse..

Answer / amith.patil17@gmail.com

Calibration Can be Carried out By Calibration Instruments
such As For Pressure Transmitter,level transmitter the
pressure pumps with mA in Series or Can be used simply
HEART with a Pressure pump,By increasing the pump pressure
the master guage shows the indication of the Standared
pressure Corresponing to which mA reading canbe Obtained
check for incrasing and decreasing Valus for 4,8,12,16mA If
Error Occurs then Set a Zero or Span Adjustment,Simillarly
For RTD,T/C in Furnace is Carried out.

Is This Answer Correct ?    16 Yes 1 No

Describe how to calibrate various types of instruments and transmitters using ISA industrial conse..

Answer / pankaj

the calibration of pressure transmitter is calibrated with
help of comparison pump or HART Protocol...

Is This Answer Correct ?    3 Yes 0 No

Describe how to calibrate various types of instruments and transmitters using ISA industrial conse..

Answer / ram

Calibrate to any instruments like pressure transmitter by using the druk calibrator to build the pressure depends on transmitter range and measure the 4-20 ma it's increase or decrease have a span and zero settings .0_30bar like initially 4ma,8ma,12ma,16,20ma.and using hart communicator also

Is This Answer Correct ?    3 Yes 0 No

Describe how to calibrate various types of instruments and transmitters using ISA industrial conse..

Answer / jadhavamit

Calibration of instruments are carried out under master equipment,they are calibrated under the "National Instrumentation Laboratory".
I can not tell u about the detail because it is under rule But i will explain it with example:

To calibrate a temperature Gauge-
The master equipment use is "Fluidless temperature Calibrate".
it work With out fuel so it is called Fluidless Temperature Calibrator the only thing u have to do is Put the End nod into the master equipment and initially set the temp. to 0degree cent. and step up to 30degree cent. with this see the reading proportionally in the Gauge no need to calibrate if it give +-0.2 reading. if it give deviation of +-5 then adjust the span and zero to set point result we been seen these is the method of calibration of temperature gauge.
Other example of master equipment are microprocessor based pressure calibrator,Macloid gauge etc.

Is This Answer Correct ?    0 Yes 3 No

Post New Answer

More Instrumentation Interview Questions

What is wet leg condition and why is it used?

0 Answers  


Hi friends, from Aug 4th onwards iam getting spikes for analog inputs in our DCS for a period of 800ms. I throughly checked every thing, finally i found that the DI, DO's from MCC panels shields are terminated in the Signal Earth where Analog Input signals shields are terminated. Is there any problem with this type of termination?. i checked the induction voltage from MCC and getting around 25VAC, please suggest me in this concern. Our plant is 8months old only. DCS is Siemens Make Simatic PCS7, and the transmitters are Yokogawa make.

1 Answers  


How to set up the various advanced process controllers, input the proper parameters, and tune those controllers given related control system information.

1 Answers   EID Parry,


Why do we use triad cable for RTD to Temp.transmitter connection? Why we don't use 3 core cable?

1 Answers   Process Control Technologies,


why vortex flow meters and electro magnet flow meters are used in water treatment plants

1 Answers   Elite Construction,






Calibration of controller (pneumatic)

0 Answers   CCIE,


How can I read the electrical drawing easily?

2 Answers   Plot Enterprise,


What is the difference between dry contact and wet contact?

2 Answers  


Where is wet leg is used?

0 Answers  


How to calculate 47Liter H2 Gas cyilnder consumtion? Cylinder inlet Pressure is 200bar and Discharge pressure is 6.0bar. Flow is 100ml/min.For the above condition how many days one cylinder can be used?

1 Answers  


What is Hysteresis mean in Instrument?

11 Answers  


what is beta ratio?

3 Answers  


Categories
  • Civil Engineering Interview Questions Civil Engineering (5085)
  • Mechanical Engineering Interview Questions Mechanical Engineering (4451)
  • Electrical Engineering Interview Questions Electrical Engineering (16632)
  • Electronics Communications Interview Questions Electronics Communications (3918)
  • Chemical Engineering Interview Questions Chemical Engineering (1095)
  • Aeronautical Engineering Interview Questions Aeronautical Engineering (239)
  • Bio Engineering Interview Questions Bio Engineering (96)
  • Metallurgy Interview Questions Metallurgy (361)
  • Industrial Engineering Interview Questions Industrial Engineering (259)
  • Instrumentation Interview Questions Instrumentation (3014)
  • Automobile Engineering Interview Questions Automobile Engineering (332)
  • Mechatronics Engineering Interview Questions Mechatronics Engineering (97)
  • Marine Engineering Interview Questions Marine Engineering (124)
  • Power Plant Engineering Interview Questions Power Plant Engineering (172)
  • Textile Engineering Interview Questions Textile Engineering (575)
  • Production Engineering Interview Questions Production Engineering (25)
  • Satellite Systems Engineering Interview Questions Satellite Systems Engineering (106)
  • Engineering AllOther Interview Questions Engineering AllOther (1379)