Describe how to calibrate various types of instruments and
transmitters using ISA industrial consensus standard
techniques.
Answers were Sorted based on User's Feedback
Answer / amith.patil17@gmail.com
Calibration Can be Carried out By Calibration Instruments
such As For Pressure Transmitter,level transmitter the
pressure pumps with mA in Series or Can be used simply
HEART with a Pressure pump,By increasing the pump pressure
the master guage shows the indication of the Standared
pressure Corresponing to which mA reading canbe Obtained
check for incrasing and decreasing Valus for 4,8,12,16mA If
Error Occurs then Set a Zero or Span Adjustment,Simillarly
For RTD,T/C in Furnace is Carried out.
| Is This Answer Correct ? | 16 Yes | 1 No |
Answer / ram
Calibrate to any instruments like pressure transmitter by using the druk calibrator to build the pressure depends on transmitter range and measure the 4-20 ma it's increase or decrease have a span and zero settings .0_30bar like initially 4ma,8ma,12ma,16,20ma.and using hart communicator also
| Is This Answer Correct ? | 3 Yes | 0 No |
Calibration of instruments are carried out under master equipment,they are calibrated under the "National Instrumentation Laboratory".
I can not tell u about the detail because it is under rule But i will explain it with example:
To calibrate a temperature Gauge-
The master equipment use is "Fluidless temperature Calibrate".
it work With out fuel so it is called Fluidless Temperature Calibrator the only thing u have to do is Put the End nod into the master equipment and initially set the temp. to 0degree cent. and step up to 30degree cent. with this see the reading proportionally in the Gauge no need to calibrate if it give +-0.2 reading. if it give deviation of +-5 then adjust the span and zero to set point result we been seen these is the method of calibration of temperature gauge.
Other example of master equipment are microprocessor based pressure calibrator,Macloid gauge etc.
| Is This Answer Correct ? | 0 Yes | 3 No |
As an industrial instrument technician what we have to know?
Draw block digram for well head from well to shipping pump?
Why use ball valves on Gas pipelines instead of gate & or globe valves...means basic diffrence or use of these valves..how designers knows this type of valve with rating is required...
0 Answers Adgas, GAIL, Oil India, ONGC,
What is the formula To convert Degree centigrade to foranhight.
What is DCS and PLC scanning time?
what is sil (safety integrity level
if conical type tank in top LEFT side solid material coming but LT MOUNTED IN TOP OF LEFT SIDE, how to measure ultrasonic level transmitter ?
suppose, after calibrating a transmitter we fixed it in a process line. At live zero condition the Tx shows 4.5 mA. what should we do to make the Tx show 4 mA? If we adjust zero, will it not affect the calibration?
what is meant by nuclear collector?
Why the Proximitor supply should be Negative voltage? Why should not we give Positive voltage to the Transducer?
Explain the difference between differial pressure & delta pressure?
what is difference between HART communication and BRAIN communication?
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)