Answer / s
When Differential pressure gauge or Transmitter put in open
to atmosphere condition,adjust zero.Apply Maximum Pressure
at HP side and Adjust span by adjusting link in the case of
Diff.Preesure gauge or adjusting span screw in case of
analog Diff.Pr.Transmitter.
| Is This Answer Correct ? | 5 Yes | 0 No |
Answer / vinesh prajapati
When Differential pressure gauge or Transmitter(H.P.
SIDE,L.P. SIDE parellal) put in open
to atmosphere condition,adjust zero.Apply Maximum Pressure
at HP side and Adjust span by adjusting link in the case of
Diff.Preesure gauge or adjusting span screw in case of
analog Diff.Pr.Transmitter.
| Is This Answer Correct ? | 2 Yes | 0 No |
DP Transmitter are used in the measurement of level/flow
measurement/DP of filters.. Here both the lp & hp points
are rather important.
To calibrate the transmitter:
confirm the cell and adapter blocks are clear,
Apply both side the equal pressure by opening the equalizer
valve (Less than its MWP-max working pressure),
should be 4 mA. (This is called Static test)
Make zero trim if found error.
Apply the required range of pressure, only in hp side and
zdj to 20 mA. Again chk 0%, if error found adj zero. After
no error found in zero/span then proceed next step.
Apply pressure and note 4-5 points of its range in hp
side., increasing and decreasing (to avoid hystersis error).
Make write protect and calibration tag on transmitter.
| Is This Answer Correct ? | 2 Yes | 0 No |
In closed loop control there are several “blocks” used to show the interactions between the process and the control system. These blocks are referred to as the “process” and the “controller.” Describe the relationship between these two blocks when creating a well tuned control loop.
please tell me five types of fault of control valve and positioners?please
what is upstream and downstream in liquid flow?
How does milli volt is created in themocouple?
whatis virtual instrumention?
whats the diff b/w zero pressure and obsolute pressure
how to change 4 wire rtd transmitter into a 3 wire rtd transmitter? as our transmitter is working for 4 wire rtd, what should do for working on 3 wire rtd transmitter
what is function of DCS , and how is it work , explain with difination
1 Answers Bapco Bahrain Petroleum Company,
what is the difference between PLC & DCS
Tell me about the synchronization panel.
when an eddy current probe is installed inside of a turbine or compressor in case of probe problem same as:crash of head of probe,changing the scale factor because of low measuring gap(physical gap is ok(example=1.14mm)but voltage is reduced(example=-6 volts),is any way to check the probe,when the ohm of probe is ok.thank you,(Karim Tavakoli)
How many input(Analog and digital) and outputs(Analog and digital) are there in DELTA v DCS Version 7.0?
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)