how to make a calliberation for deffirential pressure
Answers were Sorted based on User's Feedback
Answer / s
When Differential pressure gauge or Transmitter put in open
to atmosphere condition,adjust zero.Apply Maximum Pressure
at HP side and Adjust span by adjusting link in the case of
Diff.Preesure gauge or adjusting span screw in case of
analog Diff.Pr.Transmitter.
| Is This Answer Correct ? | 5 Yes | 0 No |
Answer / vinesh prajapati
When Differential pressure gauge or Transmitter(H.P.
SIDE,L.P. SIDE parellal) put in open
to atmosphere condition,adjust zero.Apply Maximum Pressure
at HP side and Adjust span by adjusting link in the case of
Diff.Preesure gauge or adjusting span screw in case of
analog Diff.Pr.Transmitter.
| Is This Answer Correct ? | 2 Yes | 0 No |
DP Transmitter are used in the measurement of level/flow
measurement/DP of filters.. Here both the lp & hp points
are rather important.
To calibrate the transmitter:
confirm the cell and adapter blocks are clear,
Apply both side the equal pressure by opening the equalizer
valve (Less than its MWP-max working pressure),
should be 4 mA. (This is called Static test)
Make zero trim if found error.
Apply the required range of pressure, only in hp side and
zdj to 20 mA. Again chk 0%, if error found adj zero. After
no error found in zero/span then proceed next step.
Apply pressure and note 4-5 points of its range in hp
side., increasing and decreasing (to avoid hystersis error).
Make write protect and calibration tag on transmitter.
| Is This Answer Correct ? | 2 Yes | 0 No |
how to measure temperature from resistance of rtd?and how to measure temp from voltage of thermocouple?is there any formulas?
how works encoder in plc
What is the different between of rtd and thermocouple ?
Explain, and give a real-world example of MODULATING control
Controller not working in auto mode?...what are the problem can occur...Pls explain
hi! i want to know about mpse exam.
send me last 5 years questions asked in rrb chennai, for section engineer
PT RTD Range: -200 to 500 degree celsius If I configured in DCS Lower range and Upper range . In face plate it shows red colour due to overa range how was the characteristics look If the temperature is more than operating range. how was the characteristics look . whether its in Resistance vs Temperature curve is Linear or Non linear
Can i convert 230v ac to 24v dc by using a relay?Instead of Smps is it possible? What is relay logic in plc ?? How it works?
What pressure transmitter.
what are the documents(Construction point of view for a process plant) submitted by contactor initially entering in to new site for client?
plz give me the logic for how a pushbuttoon can be worked as atoggle switch i m working with allenbradely/micrologix1200/rslogix500
Civil Engineering (5086)
Mechanical Engineering (4453)
Electrical Engineering (16638)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)