What is linearity in HPLC and how will you test?
Answer / kishor solanke (alkem)
To check the area(response) of analylite in the sample is
linear with respect to it's concentration OR injection
volume.This parameter done by two method:
1) Detector linearity:
In this linearity test the solution of different
concentration (linear conc.)are injected in triplicate of
each solution & linearity of each concentration calculated
as co-relation coefficient by graphical pattern in excle
against mean srea obtained.
2)Injector Linearity:
In this test the soluiton of same concentration injected in
triplicate with different injection (linear)volumes &
linearity calculated as above.
| Is This Answer Correct ? | 20 Yes | 0 No |
principle of UV , PDA , RI detector in HPLC
What is colimn selection Criteria in HPLC analysis?
Any molecule is filing for ANDA purpose, what are the dissolution requirements / specifications needed ?
If vendor having more imp than monograph then how to proceed? and how to set spec?
What is the difference between purge septum flow and column flow in gas chromatography?
define beer-lambertslaw.
how to qualify the impurity?
why plane of polarized light rotated by opticaly active compunds?
How to prepare 10% HCl from 36.5% HCl solution.
How can we confirm the HPLC column is end-capped or not? Is it possible to identify by physical appearance?
full name of OOS & OOT ? Dercribe it with details ? why this is needed in pharma industry ?