we calibrate HPLC once in 6 months while iam doing
calibration i came to know that calibration was failed.
then how can i justify that the samples analyzed on the
system during this 6 months was wright? how can i forward in
to the next step for those batches?
Answers were Sorted based on User's Feedback
Answer / ravi narwade
Whenever we are doing any analysis we are performing the
system suitability test for it, and if the "SST" is passing
no one will raise the question on analysis of samples on
the instrument which fails in calibration.
| Is This Answer Correct ? | 31 Yes | 8 No |
Answer / nirchana
if u hv doubt ..thn u can check previous sample tht result u known..tht means any std sample thn match with recent sample
| Is This Answer Correct ? | 2 Yes | 2 No |
Answer / k l reddy
Need not to analyze all previous batches but you should
analyze 6 previous batches(recent batches of before
calibration)on different system(which is qualified) and
compare the results with old results. Fallow the same
prociser until the resukts matched with old results.
| Is This Answer Correct ? | 2 Yes | 2 No |
Answer / sasikala
Need not to analyze all previous batches but you should
have internal sop which describes how many batches
(immediate previous batches) to be analyzed and how to be
compared. But this anwser not complete one for you.
| Is This Answer Correct ? | 1 Yes | 4 No |
Answer / pankaj punjabi
After 6 month calibration is failed,So you have to
reanalyse the previous batches which is analyse for last 6
month.and comapre the results.then only it is justified.
| Is This Answer Correct ? | 2 Yes | 16 No |
In pharmaceuticals formulation OOS applicable in Raw Material(which is not manufacturing in the company. and purchase by approved vendor)).Why?
which are the diffrent batches in the pharmaceuticals?
what is the procedure for cleaning of lenses of hatr accesory of ftir instrument?
stability study is going on up to 3years for famotidine-USP but the pharmacopeia is revised and tlc test is replaced with hplc test. At the 4th year frequency the product is failed in hplc test .how can i assain the expiry or retest date and can i stop or continue the programme?
What is polymorphism?which type of compounds exhibits polymorphism?
is ther any guidline is available for limt of peak fronting.and what is the limt of peak fronting.
What is the principle of HPLC?
I have compare C2H2-air and C2H2-N2O flame AAS on determination calcium. Both use same range of std to plot calibration curve. (2-6ppm) When i measure the sample with phosphate, KCl and LaCl, C2H2-N2O flame give false positive result, around 0.5ppm. When i measure the sample with phosphste, KCl and EDTA. C2H2-N2O flame also give 0.5ppm false positive. But both above mentioned sample would not give false positive when measured by C2H2-air flame. What is the reason?
How to Prepare Dilute Sodium hydroxide solution?
How can the GC or HPLC method is selected to determine the impurity profile in drug product?
What is the difference between purge septum flow and column flow in gas chromatography?
why 2Theta value is consider in the XRPD spectra and how to consider the 2 Theta value