we calibrate HPLC once in 6 months while iam doing
calibration i came to know that calibration was failed.
then how can i justify that the samples analyzed on the
system during this 6 months was wright? how can i forward in
to the next step for those batches?
Answers were Sorted based on User's Feedback
Answer / ravi narwade
Whenever we are doing any analysis we are performing the
system suitability test for it, and if the "SST" is passing
no one will raise the question on analysis of samples on
the instrument which fails in calibration.
| Is This Answer Correct ? | 31 Yes | 8 No |
Answer / nirchana
if u hv doubt ..thn u can check previous sample tht result u known..tht means any std sample thn match with recent sample
| Is This Answer Correct ? | 2 Yes | 2 No |
Answer / k l reddy
Need not to analyze all previous batches but you should
analyze 6 previous batches(recent batches of before
calibration)on different system(which is qualified) and
compare the results with old results. Fallow the same
prociser until the resukts matched with old results.
| Is This Answer Correct ? | 2 Yes | 2 No |
Answer / sasikala
Need not to analyze all previous batches but you should
have internal sop which describes how many batches
(immediate previous batches) to be analyzed and how to be
compared. But this anwser not complete one for you.
| Is This Answer Correct ? | 1 Yes | 4 No |
Answer / pankaj punjabi
After 6 month calibration is failed,So you have to
reanalyse the previous batches which is analyse for last 6
month.and comapre the results.then only it is justified.
| Is This Answer Correct ? | 2 Yes | 16 No |
How long we can use hygroscopic reference standard in terms of water absorb.I mean in line with respective monograph's specification or more than that?
How you can calculate pKa ?
significane of detector wavelength accrucy in calibration of hplc
If we do accuracy at same concentration at which linearity planned,what is the need to do linearity separately?
i set electrode to 7 buffer than i used it to check 4 buffer and 10 buffer if the results are not as expected than is it better to set the buffer again to 4 buffer or is there any other method available please let me know ...
In HPLC Chromatographic separation Resolution equation is Rs = 1.18 (tR2 –tR1)/wh1+Wh2, Where, tR2 and tR1 is run time of eluent ratios of the distances from the point of application to the centers of the spots and the distance travelled by the solvent front from the point of application (retardation factor), and Wh1and Wh2 , is peak width of at half height. What is 1.18 ?
what is mean by Q in S1 and S2
How precise should be the linearity in HPLC method validation?
Why dissolution apparatus having 8 jars ?
what is the relation of normality & molarity ? give me formula?
what is a differance between assay by HPLC and assay by titration?
what is principal of karl ficher instrument!