In the method validation proceedure,lod & loq calculation we use the formulae for LOD=3.3*SD/slope and for LOQ=10*SD/slope,why we use 3.3 & 10 in the formulae?
Answer Posted / veeru01
LOD :- LOD is measured 10 times independent sample of Blank
and Student-t test of 10 times, Degree of freedom 9 (n-1)at
99% confidence label is 3.25 as per Student chart. Also
Ratio of S/N =3:1
LOQ:- LOD is regarded as the lower limit for precise
quantitative Measurement . LOQ defined as concentration
which would a produce a signal 10-12 time. At this point
standard deviation is small enough (Approximately 10-15%) so
that quantified value can be deemed reliable. Also Ratio
S/N= 10:1
| Is This Answer Correct ? | 7 Yes | 3 No |
Post New Answer View All Answers
What is rs test why we are performing rs test?
Why perchloric acid used for potentiometric titration
In HPLC Calibration, On which basis RSD Limit of noise test is fixed (NMT 33.0 % )
Ratio of Water and Acetonitrile(diluent) is not recommended to use in cleaning method validation,why?
about method developement in hplc
In performance Check of GC Why Hexadecane Peak is ConsideredÂ
What is stray light in uv what are units for stray light
Which products will produce when acetic anhydride reacts with Ter-butanol, Sec-butanol, Iso butanol, Dimethylamine etc.
A ha tc is actual defination of retention time?
What is intact assay method development.....?
what are the guidelines for analytical method validations?
if tech transfer fails whay require to do?
in dissolution why pool sample needed? in which type of drug pool sample need?
how to select short coloum and long coloumn for new molecule
Why we select scan range from higer wavelength to lower wavelength in uv visible spectroscopy and ftir spectroscopy ? .