why 0.005 M sulphuric acid is used to calbrate
spectrophotometer and why caffiene is used for calibrating
hplc
Answers were Sorted based on User's Feedback
Answer / kanakadurga
A solution of 0.005M H2SO4provides ionic strength to the
solution (hence, it changes the refractive index so as to be
similar to comparable solutions), but it doesn't absorb much
in the UV or Visible region. Therefore, light hitting the
detector will be almost all stray light.
Caffine is primary standard and Shows maxima at two
different wavelength ie.205nm & 273nm and Cheap & easily
available
| Is This Answer Correct ? | 19 Yes | 0 No |
Answer / sp goswami
A 0.005M sufuric acid solution gives the required ionic
strength, required dissolution and also do not interfere in
UV-VIS Region giving 99% absorbance for the analyte of
interest so 0.005 M diluted sulfuric acid is useful .
Caffeine is a primary std ,pure form ,good absorbance and
readily available in market
Regards
| Is This Answer Correct ? | 10 Yes | 3 No |
why octanol used to determine the partition coefficient ?
What is difference b/t uv detector and rid
Which is the best mobile phase to separate organic acids using C8 column. And how should the samples be prepared for hplc?
iam using ph instrument lab india make.3ponit and 5point.in calibration time 3point slope will come 98.89and 96.87.that limit is 85 to 105.what basis slope will come.any calucalation between them.5point wiicome slope 99.89?why 3p0int came 2 values
what is use of acetonitrile compare to methanol in RP-HPLC Method devlopment
What is Formula of Potency calculation on Dried basis, As is basis& Anhydrous basis.
which brands GC coloumn available in market
why we use 0.1% acetone in gradient calibration of HPLC?
3 Answers Cipla, Dr Reddys, FDC,
give me method of calculating co relation of coefficient for HPLC,SOR & ANY OTHER
when we use pda detector in hplc instead of uv detector
We use hicrome 10 cxs column in hplc analysis....pls tell me.. what is the meaning of 10 cxs?
in which situation ion pair require to use?