today i am doing calibration of HPLC. when i inject the
blank it was going on -ve.(bleow then zero) why?
Answers were Sorted based on User's Feedback
Answer / b.gangatharan
This is due to improper sonication of mobile phase and improper stabilisation of column.
Regards,
B.Gangatharan
| Is This Answer Correct ? | 7 Yes | 1 No |
Answer / sudheer kumar m
may be mobile phase was not stable in that particular
method. so condition the column with mobile phase and re inject.
| Is This Answer Correct ? | 2 Yes | 2 No |
Negative peak simply indicate that your mobile phase
absorbs UV light.So use only HPLC grade solvents.Even with
HPLC grade solvents you get the negative peaks you system
might not be working properly.
Regards
M.KIRAN REDDY
| Is This Answer Correct ? | 2 Yes | 7 No |
Related substance method equivalency on control sample or spiked sample?
is it nessesary to do solution stability for 7 days?
what is the difference between Hydrochloric Acid & Hydrogen chloride
How to calculate the signal to noise ratio.
Difference Between End Cap HPLC Column and Non End Cap Column
why we inject sample TWICE in ASSAY by HPLC analysis but in RS once?
4 Answers Aurobindo, DRL, Neuland,
difference between HPLC & LC-MS
why use k2cr2o7 , kcl h2so4 in uv calibration
What is Formula of Potency calculation on Dried basis, As is basis& Anhydrous basis.
why we wre doing linearity siudiec in hplc method validation
in what basis wavelength will select for a compound at initial method development trials
What is UV cut off?