Golgappa.net | Golgappa.org | BagIndia.net | BodyIndia.Com | CabIndia.net | CarsBikes.net | CarsBikes.org | CashIndia.net | ConsumerIndia.net | CookingIndia.net | DataIndia.net | DealIndia.net | EmailIndia.net | FirstTablet.com | FirstTourist.com | ForsaleIndia.net | IndiaBody.Com | IndiaCab.net | IndiaCash.net | IndiaModel.net | KidForum.net | OfficeIndia.net | PaysIndia.com | RestaurantIndia.net | RestaurantsIndia.net | SaleForum.net | SellForum.net | SoldIndia.com | StarIndia.net | TomatoCab.com | TomatoCabs.com | TownIndia.com
Interested to Buy Any Domain ? << Click Here >> for more details...


acceptance criteria for ph meter, hplc
detector,pump,injector,oven when we calibrate.

Answers were Sorted based on User's Feedback



acceptance criteria for ph meter, hplc detector,pump,injector,oven when we calibrate...

Answer / guest

Acceptance Criteria
---------------------

pH Meter:
-----------

Perform 2 point or 3 point calibration as per the
use/requirement.

The slope should be 95% to 105%

Where as pH measured value for known pH buffer should be
4.01 ± 0.05
7.00 ± 0.05
9.21 ± 0.05
10.00 ± 0.05
11.00 ± 0.05

HPLC Detector:
--------------

Prepare differenct concentration of caffeine solution say
10 ppm, 25 ppm, 50 ppm, 75 ppm, 100 ppm, and 125 ppm.

Inject single injection of each solution and calculate the
correlation coefficient = sholud be 0.999

HPLC Pump:
----------

a) Flow rate accuracy
b) GPV

a) Flow rate accuracy:

Calculate flow rate accuracy using water at difference flow
rate at 0.5 mL/min , 1.0 mL/min, and 3.0 mL/min at duration
of 5 minutes and calculate the volume collected with in 5
minutes duration using stop watch collected in pre-weigh
calss A volumetric flask.Apply density of water at
specified temperature.

Acceptance Criteria: For all the flow rate should be ± 5%

b) Gradient Proportianating Value (GPV)

GPV test will be perform using 5.6 mg of Propyl paraben ----
-} 1000 mL of methanol for channel A and B if we are
calibrating the Quadranaery pump and C and D channels in
methanol run the gradient composition at the specified time
to obtain the response when compared with higest response
of methanol vs propyl paraben solution should be 10 ± 0.5
i.e 9.5 - 10.5.

We can even calibrate using 0.5% Acetone in water by
evaluating the response obtained throughout the run time.

Injector:
---------

a) Injector Linearity
b) Injector precision
c) Volume accuracy

a) Injector Linearity

Inject singe injection of 25ppm caffeine solution of
injection volume 5µ, 10µ, 20µ, 50µ and 100µ. if the
injector loop is 100 µl loop where as if the injector loop
is 200 µl then go upto 200 µl for better accuracy.

Acceptance criteria: Correlation coefficient: 0.999

b) Injector Precision:

Inject 5 times of 25 ppm caffeine solution and calculate
the RSD for area.

Accetance criteria: < 1.0%

c) Volume Accuracy:

Check the injector volume accuracy by intecting the 5µl
volume for 10 times in a pre-weigh vial containing water.

Check the injector volume accuracy by intecting the 10µl
volume for 10 times in a pre-weigh vial containing water.

Acceptance criteria: 4.95 - 5.05 µl
19.5 - 20.5 µl

Column Oven:
------------

Calibrate the column oven with calibrated
thermometer/digital at 30° C and 70°C

Acceptance Criteria: 30 ± 2 °C
70 ± 2 °C

Is This Answer Correct ?    19 Yes 1 No

acceptance criteria for ph meter, hplc detector,pump,injector,oven when we calibrate...

Answer / saptale maruti

Flow rate ± 0.2% of set value
Flow rate consistancy RSD NMT1.0%
Compositional Accuaracy : Difference in Absorbance is 0.01AU
Difference in time NMT 20SEc
Delay Volume NMT 4.5 (For 4 ml Delay Volume System)
Injection Volume Accuracy is : Average volume for 18
injection is 19.6 to 20.4µl
Linearity 0.999
Wavelenght Accuracy Maxima 273± 2 nm
Minima 205 and 245± 2 nm
Oven Temperature 2 degree from set temp.

For Ph ± 0.02

Is This Answer Correct ?    7 Yes 2 No

acceptance criteria for ph meter, hplc detector,pump,injector,oven when we calibrate...

Answer / basha

For pH + OR - 0.05
HPLC-DETECTOR + OR - 1nm
Acceptance Criteria: For all the flow rate should be ± 0.02
Acceptance criteria: Correlation coefficient: 0.98
Accetance criteria for volume:1 %
Acceptance Criteria: ± 1 °C

Is This Answer Correct ?    6 Yes 7 No

Post New Answer

More Analytical Chemistry Interview Questions

why we want incorporate area of each solution of Intermediate precision and repeatability as well as Accuracy ? What is the use to incorporate area of each solution of Intermediate precision and repeatability as well as Accuracy ?

1 Answers  


How to define samples weight in moisture content

1 Answers  


what is Beer Lambert low ?

8 Answers   MAN,


why we are using in ph calibrations ph4,7,9.2 buffers only?

5 Answers   Reddy Labs,


What is split

4 Answers   Vimta Labs,


why we inject sample TWICE in ASSAY by HPLC analysis but in RS once?

4 Answers   Aurobindo, DRL, Neuland,


various methos to calculate precision in validation methods in HPLC?

2 Answers  


mode of absorption in alimentary canal?

0 Answers   Sun Pharma,


what is concentration of oxygen in U.H.P. grade N2 cylinder

1 Answers  


How we can perform Linearity study for validation by HPLC with very low conc of sample

1 Answers  


How to calculate lamp energy in hplc?

1 Answers  


1)why use benzophenon,uracil in hplc calibration. 2)why use sucrose in polarimeter calibration. 3)what is S/N ratio, noise and drift in hplc 4)how we can find lod and loq in method validation

1 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)