In Analytical method development related substance method when did follow w/w method when did follow area normalisation method
Answers were Sorted based on User's Feedback
Answer / d.thirupathi
if the method dont have known impurities we can choose the
area normalisation method.if known impurites is present we
can go for w/w method
| Is This Answer Correct ? | 14 Yes | 3 No |
Answer / parag
we can go with area normalisation method when the RRF of
the impurities are 1 or during the Drug-Excipient
Compatibility study when the amount of drug added in the
sample not exactly known concentration.
| Is This Answer Correct ? | 4 Yes | 2 No |
Answer / tukaram masal
When response of main peak is linear from LOQ to nominal
test concentration then area normalisation method for
related substances and When response of main peak is not
linear from LOQ to nominal test concentration then w/w
method for related substances.
| Is This Answer Correct ? | 3 Yes | 3 No |
Answer / raghavender
impurites are not avaliable follo are normalization method &
other cases fololow w/w method
| Is This Answer Correct ? | 3 Yes | 4 No |
Answer / naga
when all impurities RRF values between 0.95 to 1.05 we can
go for area normalisation method
| Is This Answer Correct ? | 3 Yes | 4 No |
WHY POTASSIUM DICROMATE USING IN UV CALIBRATION
1 Answers Apotex, Micro Labs, Rene,
Why water not directed by FID detector
2 Answers Amoli Organics, Lupin,
What is the difference between the optics of UV and PDA Detectors ?
why reference flow required during hplc analysis on rid?
What is split
What is the accrptance criteria in RSD for RS method precision on basis of impurity percentages?
Why 162 mg NaOH used for 1N volumetric solution in 150 ml water then 54.5 ml diluted with 1000 ml
From where 16 or 5.545 is derived in theoretical plates calculation formula by USP or JP/BP/DAB and what is meaning of this coefficient and impact in calculation ?
2 Answers Nectar Lifesciences,
for which type of applications gas chromatography is used?
What is the lab normality??
For Recent draft guidelines for Dissolution test apparatus calibration, why removing the salicylate tables?
what is the difference b/w repeatability nd reproducibility ?