How you will select the runtime in HPLC for a drug in
method development.
Answers were Sorted based on User's Feedback
Answer / ankur dwivedi
There is no any thumb rule that the runtime should be 3
times than main peak, but also depends on the late eluting
impurities in compound; which should be confirmed at the
time of method development. Considering all these, the run
time should be selected.
Ankur Dwivedi
ZCL Chemicals, Ankleshwar
| Is This Answer Correct ? | 24 Yes | 1 No |
Answer / sarma dvv
In the method development, we have to extend the run time 7-
8 times more with analyte retion time. we have to check any
other peaks are eluting. In that run nopeaks are eluted
means accordingly we can fix the run time.
Generally for isocretic 3times more with resect to analyte
peak.
For Gradient we have to extend the organic concentration
maximum possible.
| Is This Answer Correct ? | 15 Yes | 2 No |
Answer / sundar
the runtime will be three times of main analye retention
time.
| Is This Answer Correct ? | 13 Yes | 6 No |
Answer / rahul
There are two three concerns to keep in mind when setting a
run time during method development.
1)Refer the degradation study data of the product.due to
this you come to know about the product beheviour with
method during stress conditions.
2)Confirm the impurity and main compund mass balance test
because of this you will assure about no further elution.
3)Run the HPLC for 5X to the column volume due to this you
can assure about and late elutions.
If with these experiments there is no elution after certain
time you can set the run time after 5 to 10mins after last
peak elution.
| Is This Answer Correct ? | 4 Yes | 0 No |
Answer / ajay vishwakarma
depends on the last possible impurities in compound; which
should be confirmed at the time of method development.and
degradation Study,further Run time 5.0times more W.r.t to
Analyte Peak
| Is This Answer Correct ? | 4 Yes | 2 No |
what is peak purity? how may types of methods to determine peak purity?
1 % solution is equal to how much ppm
What is the difference between Chromatograhic Purity And Related substances
Why we check moisture,ash & AIA in product
acceptance criteria for ph meter, hplc detector,pump,injector,oven when we calibrate.
why disodium tartarate is used for autotitrator calibration
In pharmacopeal solibility parameter of a material, is et essential to test all paramete? If 70-75% parameter passes can we assess that the material is satisfactory or not?
What is the principle of hplc detectors like.. UV detector,flurosence detector, RID?
Why dissolution test is not performed in all of the products
[3/30, 13:29] Manoj P Venkatpurwar: How hplc column selection according to structure? How mobile phase buffer selection on molecule structure?
how to convert normal phase hplc to reverse phase and vice versa?
what v define from IR & NMR spectroscopy ?