Answer / ankur dwivedi
There is no any thumb rule that the runtime should be 3
times than main peak, but also depends on the late eluting
impurities in compound; which should be confirmed at the
time of method development. Considering all these, the run
time should be selected.
Ankur Dwivedi
ZCL Chemicals, Ankleshwar
| Is This Answer Correct ? | 24 Yes | 1 No |
Answer / sarma dvv
In the method development, we have to extend the run time 7-
8 times more with analyte retion time. we have to check any
other peaks are eluting. In that run nopeaks are eluted
means accordingly we can fix the run time.
Generally for isocretic 3times more with resect to analyte
peak.
For Gradient we have to extend the organic concentration
maximum possible.
| Is This Answer Correct ? | 15 Yes | 2 No |
Answer / sundar
the runtime will be three times of main analye retention
time.
| Is This Answer Correct ? | 13 Yes | 6 No |
Answer / rahul
There are two three concerns to keep in mind when setting a
run time during method development.
1)Refer the degradation study data of the product.due to
this you come to know about the product beheviour with
method during stress conditions.
2)Confirm the impurity and main compund mass balance test
because of this you will assure about no further elution.
3)Run the HPLC for 5X to the column volume due to this you
can assure about and late elutions.
If with these experiments there is no elution after certain
time you can set the run time after 5 to 10mins after last
peak elution.
| Is This Answer Correct ? | 4 Yes | 0 No |
Answer / ajay vishwakarma
depends on the last possible impurities in compound; which
should be confirmed at the time of method development.and
degradation Study,further Run time 5.0times more W.r.t to
Analyte Peak
| Is This Answer Correct ? | 4 Yes | 2 No |
which light is used in FT-IR spectroscopy
What is difference between" LOD" AND "VACCUME LOD"
what is mean by ambient temperature?
In rs method development when we are going area normalization method to dilute standard method?
Which type of column should i use to check the purity of high molecular weight protein using HPLC reverse phase column chromatography? Hi everyone. I wanna to check the purity of high molecular weight protein (collagen) with MW of ~130 kDa using a HPLC. I know C18
0 Answers Petronas, Shanghai Ocean University,
When i inject amines and acid on HPLC then which peak will elut first?
Why are use silicon oil mr?
What is unit of KF?
What is oot in Stability analysis? (example)
What is the limit of minimum area in in hplc RS anlysis
what is the film thickness in ftir
what is MCG dilution in HPLC & GC analysis
1 Answers Amoli Organics, Cipla,