How you will select the runtime in HPLC for a drug in
method development.
Answers were Sorted based on User's Feedback
Answer / ankur dwivedi
There is no any thumb rule that the runtime should be 3
times than main peak, but also depends on the late eluting
impurities in compound; which should be confirmed at the
time of method development. Considering all these, the run
time should be selected.
Ankur Dwivedi
ZCL Chemicals, Ankleshwar
Is This Answer Correct ? | 24 Yes | 1 No |
Answer / sarma dvv
In the method development, we have to extend the run time 7-
8 times more with analyte retion time. we have to check any
other peaks are eluting. In that run nopeaks are eluted
means accordingly we can fix the run time.
Generally for isocretic 3times more with resect to analyte
peak.
For Gradient we have to extend the organic concentration
maximum possible.
Is This Answer Correct ? | 15 Yes | 2 No |
Answer / sundar
the runtime will be three times of main analye retention
time.
Is This Answer Correct ? | 13 Yes | 6 No |
Answer / rahul
There are two three concerns to keep in mind when setting a
run time during method development.
1)Refer the degradation study data of the product.due to
this you come to know about the product beheviour with
method during stress conditions.
2)Confirm the impurity and main compund mass balance test
because of this you will assure about no further elution.
3)Run the HPLC for 5X to the column volume due to this you
can assure about and late elutions.
If with these experiments there is no elution after certain
time you can set the run time after 5 to 10mins after last
peak elution.
Is This Answer Correct ? | 4 Yes | 0 No |
Answer / ajay vishwakarma
depends on the last possible impurities in compound; which
should be confirmed at the time of method development.and
degradation Study,further Run time 5.0times more W.r.t to
Analyte Peak
Is This Answer Correct ? | 4 Yes | 2 No |
Ratio of Water and Acetonitrile(diluent) is not recommended to use in cleaning method validation,why?
0 Answers Aurobindo, Reddy Labs,
What is the principle of Thermal conductivity detector and FID?
in which situation require to use paddle and basket?
significane of detector wavelength accrucy in calibration of hplc
How do you show that equal quanitities of two isomers provide the same response using a PDA
What is control room temperature and which guide line says?
For Recent draft guidelines for Dissolution test apparatus calibration, why removing the salicylate tables?
What is linearity in HPLC and how will you test?
For Dissolution test why we are Performing 6 Bowls why not more.any were mentioned or in what basis we are performing? --------Veerabhadrarao.M
2 Answers Lee Pharma, Micro Labs,
in uv visible spectroscopy calibration why we are using potasium dichromate and tolune
which situation gc hs and gc als require to use?
What is impurities in pharma analysis?